fume hood airflow is cooling off hotplate


I have a good hot plate and whenever I use it inside my fume hood, it takes FOREVER to heat up. Whenever I start the blower after my oil bath is heated up, it cools everything down. The face velocity is right around 100ft/min (measured with a Vaneometer) so decreasing it substantially isn't really an option, otherwise it defeats the entire point of a hood.

Is this a common problem? All I can find when searching is 'malfunctioning hot plate'. Should I just accept that I'll have to run it +25°-50° hotter and dial in the 'actual' temperature (like calibrating a thermometer at non-sea level by boiling water) than I would normally, or is there something in my setup that I can change? My hotplate is on a scissor lift about 5 inches off the bottom of the hood.
View user's profile View All Posts By User
NZniceguy
Harmless
*




Posts: 48
Registered: 16-9-2017
Member Is Offline

Mood: No Mood

[*] posted on 9-8-2019 at 18:54


Maybe put a shroud around the hotplate?
View user's profile View All Posts By User
draculic acid69
International Hazard
*****




Posts: 1371
Registered: 2-8-2018
Member Is Offline


[*] posted on 9-8-2019 at 21:08


Definitely try the shroud idea.its the simplest fixed
View user's profile View All Posts By User
Ubya
International Hazard
*****




Posts: 1231
Registered: 23-11-2017
Location: Rome-Italy
Member Is Offline

Mood: I'm a maddo scientisto!!!

[*] posted on 9-8-2019 at 21:48


100ft/min is around 50cm/s, if i remember correctly tge minimum was 10cm/s and a good value was 30cm/s, so maybe you could dial down a bit your hood blower.
at uni i always worked under hood, and i don't remember to have ever had a problem with a hotplate being cooled too much by the air flow





---------------------------------------------------------------------
feel free to correct my grammar, or any mistakes i make
---------------------------------------------------------------------
View user's profile View All Posts By User
CouchHatter
Hazard to Others
***




Posts: 146
Registered: 28-10-2017
Location: yes
Member Is Offline


[*] posted on 10-8-2019 at 09:57


Valuable information, all of the above. I guess I'll have to retrofit a belt pulley system for my motor to slow it down. It's a 4-speed motor on speed 1. I'll try the shroud first, but was planning a pulley anyways for explosion-proofing the blower.
View user's profile View All Posts By User
DavidJR
National Hazard
****




Posts: 908
Registered: 1-1-2018
Location: Scotland
Member Is Offline

Mood: Tired

[*] posted on 10-8-2019 at 10:37


Face velocity is a terrible performance metric for fume hoods. It has very little correlation to the ability of a hood to actually contain internally-generated fumes. In fact, increased turbulence resulting from too high a face velocity very often results in much poorer fume containment than would be achieved with the same hood if the flow rate was decreased.
View user's profile View All Posts By User
Dr.Bob
International Hazard
*****




Posts: 2656
Registered: 26-1-2011
Location: USA - NC
Member Is Offline

Mood: No Mood

[*] posted on 11-8-2019 at 16:52


I have seen the same issue when running reactions without good temperature control, when the hood is one full, the reaction stay at a one temp, but if it is on standby mode (very low flow), the temp can climb quite higher. That can actually boil reactions dry at night when the hood go to low mode.
View user's profile View All Posts By User
Tsjerk
International Hazard
*****




Posts: 3022
Registered: 20-4-2005
Location: Netherlands
Member Is Online

Mood: Mood

[*] posted on 12-8-2019 at 09:14


I use aluminium foil for high boilers, works great.