Sciencemadness Discussion Board

Thoughts On Anodes

 Pages:  1  ..  10    12

woelen - 11-2-2015 at 08:14

Your anode is a bad anode, probably not meant to be used in concentrated solutions. Swimming pool anodes are used in very very dilute solutions. They probably wear out in such solutions as well, but because of the very low rate, they last for a long time when used as swimming pool chlorinator. In your situation, you are using the anode in a much more concentrated solution and they wear out at a 100 or so times as high rate as in a swimming pool.

My MMO anodes have been used for many days in concentrated brine at 5 V and I did not even see a tiny little speck of degradation and the liquid remained clear as water, no flocculent stuff nor junk floating on the surface. With bromide I had degradation, but not of the MMO, but of the titanium substrate.

papaya - 11-2-2015 at 08:21

Thanks woelen, never thought there are different kind of MMOs, especially when it was stated as "Ruthenium/iridium oxide coating", very interesting.

Fulmen - 11-2-2015 at 08:45

So how do one pick a suitable one?

papaya - 12-2-2015 at 08:58

Some interesting news regarding my "bad" anode I thought I have to share with you. As I already said the problem was the anode seemingly being destructed during NaCL electrolysis even at low current densities - after some time I evidenced brown junk floating on top of liquid and other white stuff in the volume, several attempts -all with the same outcome.

But before discarding the anode as a "junk" one I decided to test if changing NaCl "brand" may help since we cannot be completely sure on table salt's real composition! Guess what ? I bought salt from other supplier - the cheapest brand of almost technical quality, even containing some earth as it occurred (it's just halite mineral ground down without purification). This time NO brown junk during electrolysis - almost clear solution, and only after 12 hours I discovered some little amount of white stuff floating in the volume. I'm still thinking what this could be - maybe some calcium or magnesium salts turning into insoluble hydroxides?
Can anyone tell me how to purify NaCL for further tests ? Recrystallization may not work since temperature dependance of NaCl solubility is weak, is there any other option?
Hell, table salt I used previously might contain some additives at ppm levels (perhaps iodine and ferrocyanide), how this could affect results so much?

Hawkguy - 12-2-2015 at 09:02

Platinum anodes will tarnish and degrade in extremely alkaline solutions. Better use an MMO anode.

papaya - 12-2-2015 at 09:05

That is about MMO!

Hawkguy - 12-2-2015 at 09:32

To purify NaCl just recrystallize it, filtering off insolubles while the NaCl is in solution. You could also try washing it with solvents I guess..

papaya - 12-2-2015 at 09:40

How do I recrystallize if solubility doesn't change with temperature? Really folks, how to purify NaCl?

Varmint - 12-2-2015 at 09:48

"Table salt" is probably not the best choice. First, while safe, it is not pure BY DESIGN, it has iodide, and more importantly (in terms of mass of offending product) anti-caking agents.

Your best bet is water softener salt, and if your ultimate goal is potassium chlorate or perchlorate, just bite the bullet and start with the potassium chloride.

I'll be the first to acknowledge how nifty it is to keep the NaCl concentration up, knowing the liquor is slowly being saturated by the much more soluable chlorate, but once you've done the "double-decomp" and see the KClO3 falling out of solution, it's pretty much all you need to see, especially when you consider you had to buy the KCl anyway.


papaya - 12-2-2015 at 10:04

Varmint, I don't want to perform chlorate synthesis, I'm just testing "suspicious" MMO anode I was given. Suspicious due to the problems I described, one of which(brown junk) is solved now. Now I suspect "hard" cations in water/salt, thus I ask for a method of purification.

Fulmen - 12-2-2015 at 10:13

Good question. A wash would remove any readily soluble compounds from the surface, that could be enough to get rid of caking agents. As for recrystallizing I suspect the only practical route would be to dissolve it and boil it down again. Or can one utilize a mixed solvent system?

Varmint - 12-2-2015 at 10:54

The reason for my comment was, I have MMO from the same supplier, and there are no such unexpected results in my cells.

My cathodes are Titanium grade 2 (chemically pure), the anodes are attached to similarly pure titanium rods through a special process I developed myself.

I've run moderate current density, and I've cranked the current up to expected point of failure and never saw anything but what I would call "perfect operation".

If you have MMO, you either plan to make chlorates, or, if you use a salt bridge or expensive commercial membrane, hydroxides and hypochlorites.

If you bought it "to test it", it is your test that needs work, not the quite excellent anodes.

papaya - 12-2-2015 at 11:05

I don't remember where I mentioned seller's name, whom you defend so much Varmint? And please don't start giving me "advices" what I should do with anode and what I shouldn't, since the test I described is far from being damaging to this type of anodes and the results are "no more than catastrophe" so far. Of course I could openly blame the seller for the shit he sells sometimes (maybe depending who is the buyer and from where, ah?) as well as not answering the messages I sent him, but I don't do that unless I can be 100% sure. So why are you 100% sure anodes are "quite excellent" ?

Varmint - 12-2-2015 at 14:21

I'm not defending Laser so much as pointing out what a moron you are for using tap water and table salt to "test" MMO anodes you bought to make chlorates.

Clear enough?

And I'm quite sure the anodes are excellent because used in a scientifically competent manner, they work as designed without the slightest deviation from expectation.

papaya - 12-2-2015 at 14:36

Now explain me in a scientifically competent manner how did you purify NaCl to get super results, since I don't remember anybody writing here they used analytical grade and deionized water to avoid some strange flocculent precipitate (which I'm afraid is TiO2). That was my original question, which you are now trying to turn into flame! moron? that's how you are treated by everybody maybe, unless you get more respectful to other people.

Varmint - 12-2-2015 at 15:05

I'll remind you to look at my original post, to which you replied "an anode I was given", yet earlier you indicated you obtained it from "the well known seller" (who's name happens to appear on multiple pages here).

So, you bought it, didn't have a clue how to perform even a basic test, and are now claiming to be researching this probably sub-standard anode. Hint: Do not change your story mid-stream as a means of preserving your integrity.

Then you amped up your response to include me not giving you advices, his anodes being shit, meanwhile there are lots of buyers here that are quietly churning out all manner of chlorates, hydroxides, and hypochlorites because, once again, the anodes work well when put to use by those qualified to understand how they work and how best to put them in service.

You try to pass yourself off as a qualified chemist, yet you used tap water and table salt? Have you never seen how much gunk is in potable water that has no place in a scientific procedure? Especially one where you pretend to be characterizing performance? You claim your original question was how to purify? Look back to your first post in this thread, then apologize for trying to lie to me.

You shouldn't have posted in a thoughts about anodes thread, you should have opened one of your own, in beginings, stating "I bought an anode, now what?"


papaya - 12-2-2015 at 21:59

What? You imply I'm hiding something because I said "given" instead of "bought" this time? If I had to hide something would I ever post here about issue? listen, I'm getting tired of this pointless dialogue with you honestly and don't see what are you try to prove. I told the story as it happened (and as it progressed later) and finally asked for a method of purification of NaCl, since only then I can make any conclusion! Yes I used tap water as many others did with NO problems, isn't it? What is the meaning of using distilled water if you have to use technical/table salt?
Or was it supposed that I will act like many clueless people, who bought anodes and just hurrily put it into 100amp chlorate cells directly with KCl,so nobody sees strange precipitate mixed with tons of chlorate? I invested into anode mostly because it's a valuable and not so easy to get thing where I live and for broad purposes ,future experimentation, for chlorates I made my own cobalt oxide anodes one time btw and got tired of "chlorates"(as you continually retiterate) already back in the past. Why is it "suspicious" when one tries to look more thoroughly into a thing for which he paid, especially if there're reason to be worried?
Anyway, I don't like this kind of dialogue, If anybody wants to add on the subject only then I'll answer, this is lost of a time.

violet sin - 12-2-2015 at 23:29

hey there guys,... I'll just cut in here for a sec, I bought laser's MMO as well. in addition to the couple I spot welded( shown bottom page 44 this thread) I made a larger single plate cathode from Ebay sheet stock, gr2 Ti, in order to have a larger surface area. dremel action to liberate it from a larger sheet. looked great, all shiny, polished and blunted edges,. for about 2 nights... ya I know, so what a cathode, boring, well..

I had been making bleaching solution for cleaning up, not chlorate,.. I have a pool that I hate getting green. it was measured by occasionally stirring and placing one drop on a cut paper towel tube. brown, free and convenient; showed the bleaching action very well. any how I was using both morton iodized and solar salt( softner), in a small jam jar as an initial test of the spot welds I had made in open air with out any flux. they were fine and lasted 5+ days, on while I was home, with out so much as a blemish. the solutions were clear but became strong yellow, with out ever turning cloudy. when I scaled up with the new shiny hand made one. we had an empty gallon wine jug and I wanted to make bleach/bleaching solution in it. after working everything out in one evening, I was in a rush to get sleep before too late... and wired them electrodes backwards! plugged it it, set it to 5V 1.5A awoke in the morning to a wonderful mess. the gallon wine jug was half full of a white voluminous chunky stuff that shifted to a more brick red and much thicker foam at the top of the jug. I was horrified and immediately figured what I had done. called my self a dork and went to work bummed.

later on inspection the MMO was covered in red/orange/white foamy crap and the (intended)cathode plate had been cut down by more than a third. albeit, the erosion looked interesting( bright side of things right?) the surface was still shiny were it wasn't eaten, all the revealed metal was black colored. I had assumed both were trash. but after a cleaning and changing purpose,the MMO is a nice dark black with shiny cut edges. no damage. I had been using it as counter electrode to plate copper out of E-waste I have, onto copper tube and sheet. I swear the thing is bullet proof, for a lill MMO strip I spot welded my self.

I have had the thing running with e-waste( HCl + cellphone and computer boards), sulfamic acidified e-waste, sulfamic + cupronickel, strong presence of HCl, sea salt, Iodized table salt, and countless hitch hikers from the trace resists/enamels/solders/inks/wide array of metal salts in sol. not all at one time, but many of the parameters overlapped in at least a few experiments. its still running like a champ.

the small scale tests with the e-waste, was attempting systematic depletion of specific metals in sol. none has ever been dumped and I am getting tired of saving batches to deal with. boil down( out side and slowly, neutralize, drop in aluminum scrap and store the cake. things looked good, copper was plating out easily and made some impressive yet feathery( brittle) crystals. A touch of citric acid and sulfamic seems to smooth the process a bit. I only noticed a darkening of the crystals once, but it was oxidation as far as I could tell, nothing else was plating with it at least. when I changed from the portable electrolysis device I had made, to an ebay bench top power supply, I accidentally switched the polarity and partially copper plated the MMO, then switched back and stripped it off. no harm yet again.

some of this could easily be considered just dicking around, and it was other than the first hand learning and subsequent reading, notes and what not. regardless, the MMO is far more resilient than I would have thought.

Varmint - 13-2-2015 at 02:52

papaya:

Thanks for finally coming clean and getting your story straight or straighter, I'm sure you had to go back and read what you wrote to get it "right" this time. Too bad you weren't man enough to apologize though.

See, I can amp up the "ass" with the best of them, and if you had shown even a hint of respect along the way, you would have found I can amp up the "good guy" with the best of them too.

My advice for you is to back up a bit, think before you post, and try, as hard as it might seem at first, try not pass yourself off as a research chemist, especially when peoples posts based on the evolving truth directed you to ask such an elementary question as how to purify a salt.

Have a great day bud.

mysteriusbhoice - 27-1-2016 at 08:39

I am new to chlorate synthesis using mmo and i have read that tap water is bad for your anodes but the place where i live in has tap water from desalination plants with almost no fluoride content and so far i do not see any erosion and the solution remains clear after 2 weeks and 2 runs also the anode is still rough and not deep black and smooth.

But i heard that iron and barium is really bad for them only time will tell if i had made a huge mistake.

I couldn't find any bottled water brands that are any safer and spending on an RO DI system just for this.

Detonationology - 27-1-2016 at 18:18

Quote: Originally posted by mysteriusbhoice  
so far i do not see any erosion and the solution remains clear after 2 weeks and 2 runs also the anode is still rough and not deep black and smooth.

How much power are you using?

yobbo II - 29-9-2016 at 12:19


Has anyone ever suggested using Pt leaf wrapped onto a titanium rod for an anode.

Pt leaf is available here

http://www.goldleafsupplies.co.uk/acatalog/Platinum_Loose_Le...

Platinum is 21.45g/cm^3
1000 sheets of 8 by 8 cm weighs 32 grams. I make the thickness of the sheets 2.33 micron. Too thin I would imagine.

halogen - 29-11-2016 at 13:14

https://books.google.com/books?id=x-oOAAAAYAAJ&pg=PA71&a...

Chromium phosphide was "not seriously attacked" by chlorine gas evolved upon it when used as a material for an anode. Not sure about the specific conditions in a chlorine cell (pH?) as opposed to the one from which this observation was made -- which might be unreliable as it was made in 1892. "When, in 1892, columbus sailed the skies blue" as they say.


[Edited on 29-11-2016 by halogen]

yobbo II - 3-5-2017 at 14:39

Glassy carbon rods are available here

http://onyxmet.com/index.php?route=product/search&filter...

Cheap is anyone want to have a go...

The yob


Bi-Electrode from Platinum and Titanium

yobbo II - 4-2-2018 at 16:51

Running a platinum anode for some time now making chlorate and perchlorate. It's best to just make perchlorate only as the anode is very small and slow to make chlorate. It makes 150 grams sodium perchlorate per day from sodium chlorate at three amps and is made from a flattened Pt one gram pamp suisse bullion bar 'nipped' between two pieces of Titanium (like a pair of pincers).
There is a similar set up in the paper attached.
More details to follow.

Yob

Attachment: PrBiElectrode1.pdf (233kB)
This file has been downloaded 554 times IMG_0641.jpg - 3.4MB

[Edited on 5-2-2018 by yobbo II]

yobbo II - 12-3-2018 at 19:16


I have been running the above anode for some weeks now. Works perfect.
I hammered the platinum bar so that it has twice the area of its original area and current is run at 3 amps. This makes about 130 grams perchlorate from chlorate per day.

I hope to substitute the Ti with W welding electodes as the W is more OTC. Don't know if it will work or not.
Details below:
Some pictures attached.

Platinum Anode from 1g bullion bar


On the board 'AmateurPyro.com' there was described by a poster, PDFBDQ, a way to attach MMO to a piece of Titanium so that the MMO would be completely submerged and therefor no MMO area being used as a current runner and wasting good MMO real estate! This connection was a rivet make from Titanium and worked well, even though the connection was rather sloppy, so long as the connection was underneath the surface of the liquid in the chlorate cell. Having seen this it was decided to try a similar connection between a piece of Platinum bullion and a flat piece of Titanium as a current runner. This allows all of the Platinum to be submerged and therefor almost all of its surface area to be utilized for chlorate or/and perchlorate production.
It would be unrealistic to use a small Pt bullion bar in a cell and expect some of the Pt to be used as a current runner from the anode connection (whatever metal that may be) and the surface of the cell liquid. Most of the bar would be used up and still the connection would be exposed spray/gasses etc. You would need a piece of Pt a few centimeter$ lon£.
A similar electrode is also called the "bi-electrode" as described in Platinum Metals Review 1960, 4 (1) Ti-Pt and Ti-Pt Bi-Electrodes.
A one gram PAMP Suisse bar of Platinum was obtained. It measures 14.7 by 8.9 by 0.36mm, giving a surface area of approx. 2.6cm squared. At one Amp this give a current density on the anode of 385 mA per square cm. which is a reasonable CD. I have seen Platinum run at higher CD's. This gives around 44 grams of Na perchlorate per day (from chlorate) at around 80% CE.
A constant current supply was used for all (one and three Amp) runs.



The Platinum bar is held in a bent piece of flat Ti by a Ti bolt. The cathode is held in place by bolting it to the anode current runner using a piece of perspex as in insulator and spacer which is not visible in the picture. The bolt screws into a hole in the Ti which was threaded using a simple threader.
The cathode in the vicinity of the anode is like a large blunt fishing hook. All Titanium including the bolt is grade one. Grades 2, 3 or 4 would be ok. Grade 5 which contains Aluminium would likely corrode. The weigh of the piece of Platinum was weighed at 0.98 grams using a cheap, two decimal point scales. I guess the scales was out by 0.02 grams?

The electrode was put to the test in a 'jam jar' cell containing dissolved sodium perchlorate (no chlorate or chloride) at one Amp for 15 days. Cell temperature was around 30C and cell Voltage around 6 Volts. There were no noticeable problems or weight loss from the Platinum.
Brown deposit on Pt in the picture is from just after a run in a K perchlorate cell (see below) BTW.

New cell liquid which contained 424 grams sodium chlorate per litre of solution was added to the cell. No chloride. Cell volume was around 360ml giving a total weight of sodium chlorate of around 153 grams. Current efficiency was around 75% when chlorate concentration was sensibly high. Five days later chlorate concentration was measured at 0.5 grams per liter of solution. Platinum seems to be very good at rapidly reducing the chlorate concentration to very low levels. No noticeable weight loss on Pt bar.
The anode makes around 44 grams of Na perchlorate from chlorate per day.

Another cell was started up with 500g/l Na chlorate and run for about 5.5 days. Chlorate at the end of run was measured at 0.08 grams per litre. There was a smell of ozone from the cell at the end of run. No weight loss for the Platinum. The cell pH was not measured.


Potassium Perchlorate cell
Another cell run was performed this time using potassium chlorate as starting material at 168 grams per litre. The cell was kept warm at around 60C using a hot plate as potassium chlorate is not very soluble when temperature is low. CE was lower than Na cells at around 50 to 60%. The formed perchlorate was inclined to accumulate on the anode/cathode electrode and had to be brushed off two or three times.
The cell was run to a point where there would be very little chlorate left in the cell liquor. Serious corrosion of the Titanium current runner going to the Platinum bar occured just below the water line and this caused a deposit of grey Ti compound in the cell as the end of run approached. Something to be watched if running K perchlorate cells but this problem did not occur on the next cell run. The actual cause of this corrosion is unknown. See picture. BTW the Ti in question was not 'heat damaged' as in strontium cell below.
The Pt bar was not effected.


Another K cell was started up similar to the last one. 161 g/l K chlorate. Cell volume 372ml (60g K chlorate total). 60 hours later the cell was stopped. Chlorate (in liquid) was around 1.6 grams per litre. Perchlorate was on the cell bottom. The was a brown deposit, see picture at top, on the Pt which cold oxalic acid did not remove. The serial number on the Pt was getting hard to read and the surface of the Pt was getting a 'scrubbed' look. There was no measurable weight loss on Pt bar with a cheap 2 decimal point scale.


Lithium Cell
Another cell was started using Lithium chloride as the starting material. The chloride was obtained from ceramic grade carbonate + HCl acid. 96.5 grams of Li chloride was dissolved into the cell giving around 261g/l in a 360cc volume. Current was one Amp using a constant current supply. A trace of perchlorate showed up in the cell after some hours but only a trace. Twenty five days later perchlorate stared to form as the chloride level went down to around 25 grams per litre. The chlorate run give around 50% CE. There was no attempt made to control pH. Chlorate making with a small anode like this is very slow. Seven days later the cell was shut down. Chlorate level was measured at around 0.03 grams per litre and chloride level was measured at less than 0.36 grams per litre. The cell contents was heated to around 150C for 6 hours to obtain the anhydrous perchlorate salt which (I believe) in not possible to get without vacuum, so I probably had a mixture of the trihydrate and the anhydrous. Lithium perchlorate is the only perchlorate with a known definite melting point. It does not decompose at its melting point.
Total weight of the perchlorate when it went solid was 217 grams. There was no weight loss on the Pt, it still weighed 0.98 grams.
Modified anode
It was decided to attempt to make the anode larger by hammerng the Pt bar and making it thinner. Pt leaf is made professional by placing the Pt between layers of velum (light leather) and hammering. Two pieces of ordinary leather were obtained and an approx. 250g hammer with one side flat was obtained and hammering began on a solid steel slab with the Pt between the two pieces of leather. The Pt flattened easily enough, its area was increased to twice what is was. The area of the Pt was now 2.62cm squared per side. The anode/cathode assembley was simplified with the platinum now being clamped between two flat pieces of Ti not unlike a pair of pincers with a very broad nose. The Ti bolt (grade one) together with a Ti washer, the Ti with the threaded hole and the Ti current runner was used to do the clamping. Current was run at three Amps, which is a cd of 575 mA per square cm, quite high. The same cathode was used as before. This 3 Amp anode will give around 130 grams sodium perchlorate (from chlorate) per day at 80% CE.


Strontium Cell

A Strontium chlorate/perchlorate cell was set up using Strontium chloride obtained by adding 276 grams pottery grade Sr carbonate to HCl. There was approx. 2 grams insoluble stuff which was filtered out giving a clear solution of 296.3 grams chloride which is 1.87 moles (calculated from carbonate). The purity of carbonate from the ceramics store is not known. It is best to leave some unreacted carbonate behind as metal impurities are inclined to stay behind when this is done. A new 'jam jar' was obtained which contained 770ml solution giving 385 grams chloride per litre.
The anode was run at three Amps using a constant current supply for approx. 15 days. There was a problem with (I presume) insoluble strontium hydroxide building up on the cathode which caused the cell voltage to rise and the current to go to zero. A stirrer bar about one inch long and a magnetic stirrer at 800rmp was used to stop this happening. The temperature of the cell was around 38 centrigrade. pH was controlled be adding 1 or 2 cc of 12% HCl acid to the cell per day. This is very little acid compared to a sodium cell. pH was measured with a pH probe that was calibrated with a buffer solution. Cell pH stayed around 6.8 for the chlorate stage and was at 1.8 at the run end. Acid was not added for the last few days.
The cell ran in the region of 38C. Cell voltage was inclined to move around a bit and varied from 5.5 to 7.5 Volts during the chlorate stage. The cell voltage would start to rise immediately after the stirrer was stopped.
Perchlorate appeared in the cell soon after it was started. I did not measure how much but it was a significant amount judging by the addition of methylene blue to a cell sample. 156 hours later (6.5 days) or 468 Amp-hours later perchlorate started to form in earnest. Chloride concentration was around 200 g/l at this stage, seems very high.
Acid still needed to be added to cell for 5 more days to keep pH around 6.8. At this stage (day 12) there was a very 'mild' smell coming from the cell which indicates the chlorate stage is over. Chloride concentration on day 12 was 44g/l with 7.7 Volts accross cell and a pH of 6.7.
At the start of the run there was a ss bolt holding the cathode just above water line which had corroded and discoloured the solution. This discolouration had now deposited onto the side of the jar leaving the liquid clear. It was decided to empty the container and wash the deposit off the jar inside. This was done using a tooth brush and the anode/cathode was also washed and scrubbed with the tooth brush. When the cell was started again with the stirrer as before the voltage accross the cell started to rise and the current went to zero. The rpm on the stirrer had to be increased to 1100rmp to keep cell current going at 8.6 volts. The voltage accross the cell reduced after a while and the stirrer was lowered back to 800rmp. I think this problem was caused by the cell cooling down and when it got heated up again it ran in its usual fashion as something similar happened a few days later. The solubility of the strontium hydroxide is greater in the warm cell and, I guess, builds up on the cathode less.
On day 12 the cell contents went grey as there was Titanium corrosion happening. It happened on the current runner as outlined in the picture. It was not the usuall pitting corrosion that occures but a much larger area of the Ti was etched. pH of the cell was measured at 6.2. This piece of Ti (Grade one) was subjected to high heat sometime in its past that made it very brittle. When placed in a vice and struck with a hammer it would snap as opposed to bend. Perhaps its corrosion resistance was compromised? The other pieces of Ti (not high heat damaged) did not corrode. The cell at this point had sizable concentrations of both chloride and chlorate which is a potential region for Pt to erode, and perhaps Ti too! (as described elsewhere).
The cell was stopped, contents filtered and the anode/cathode washed and the cell started again. There was no more serious corrosion problems. The cell had to be warmed up using the hot plate to stop the voltage rising too much and causing the current to go to zero even at 1100 rpm (up from 800). The cell ran OK but it was noticed that the rpm of the stirrer had to be kept at the increased rate of 1100 rmp to stop the voltage from rising too much accross the cell. This problem stopped when the cell warmed up (hydroxide has greater solubility?). The actual junction between the Ti and Pt was not the problem because when anode and cathode where shorted together the current ran at 3 Amps.
Day 13, chlorate was measured at 161 g/l. Day 14, chlorate was measured at 56g/l and the chloride was at 12g/l.
The solution was getting a bit cloudy (corrosion?) with the pH at 1.8. The voltage accross the cell was 8 volts with the cell at 38C and decreased to 6.4 volts When the cell was warmed to 56C using the hot plate (cc of 3 Amps and 1100 rpm). The end of day 15 saw chloride at 4.7g/l and chlorate at 0.3g/l. The chloride seems to be higher that the chlorate? There is small amount of white Ti corrosion compound visible between the Ti and the perspex piece holding anode and cathode apart but nothing too serious.
The Pt piece was removed and weight at 0.97 grams, down from 0.98 grams. Some Pt erosion it would seem. It is hard to say as the scales I have are just too crude for the job. A three or four place scale would be required. When the scales was taken to a warmer place it was inclined to read 0.98 again.


The cell was let settle for three days to allow cloudyness to deposit on cell bottom. Clear liquid was decanted off and the rest filtered. Liquid was dried off by placing the cell contents in a flat glass tray in an oven and heating to 170C for about 24 hours. It was very difficult to get rid of the water. A total of 474 grams (1.65 moles) of Sr perchlorate was obtained, there should be 535g (1.87 moles)). Some grams product would have been lost when doing titrations etc. The amount of product does not correspond to the amount of starting Sr carbonate. There is approx. 60 grams missing? I have no idea where it could have gone or what error was made. Total Amp-hours into cell was 1049Ah which is 39.14 moles of electrons. Having 1.65 moles of product gives 1.65 x 2 x 8 = 14.952 moles of electrons needed, giving a current efficiency of 56.6 percent.

A larger cathode could be used to stop the low solubiltiy of strontium hydroxide causing cell voltage problems.
To end Ti corrosion problems a more 'valve' valve metal could be used like Niobium. This will cost more but if the cell is being stirred the anode only needs to go below the surface of the electrolyte and therefor you would need very little valve metal.

Barium Cell
Note that barium is toxic. Barium perchlorate can be useful for making perchloric acid and other perchlorate salts as the sulphate is insoluble. To make perchloric acid you just add sulphuric acid and remove insoluble Barium sulphate.
A barium chlorate/perchlorate cell was started up. Chloride was obtained from 500g ceramics store carbonate + HCl acid. Not all of the carbonate would dissolve with perhaps 10% left over. The mixture was left overnight to get as much to dissolve as as possible. Perhaps heating might have helped? The purity of the carbonate is unknown. HCl was 12% concrete cleaner (no additives). It is best to leave some unreacted carbonate in the system as this has the effect of removing metal impurities.
The cell and anode/cathode was the same as the last cell above. Current was 3 Amps, constant current supply, with V at 5.76 and temperature at 38C. Stirring was not required but was used anyways. The solubility of barium hydroxide is higher that strontium hydroxide. pH control was attempted and about 2cc 12% HCl was added per day to keep pH around 6.2 during chlorate stage. The pH of the cell at shut down was around 2. The cell was cloudy from the start due to some carbonate that was not properly removed and turned a nasty brown colour after a few days. The cell was shut down for two days to allow the cloudyness to settle and the clear liquid was decanted off.
Cell restarted and was run until chlorate concentration was 0.76g/l. There was a black discolouration in the cell at the end of the run together (after black stuff had settles) with a very slight haze of gray (titanium erosion product?). The cell was let settle for about a week as it took that long for the gray to settle to the bottom and the clear liquid was decanted off. The solution was placed in a flat glass baking tray in an oven and dried at 130C for about six hours. It was much easier to dry than the Sr perchlorate above. Total weight of product was 533g or 1.585 moles.
Total Amp-hours into cell was 1191Ah or 44.44 moles of electrons.
The current efficiency (ce) was 57%. Final weight of Platinum bar was now 0.95, down from 0.97 grams so there was some loss of Platinum. Thats 37.5 grams of Platinum per ton of perchlorate produced. I have seen figures for platinum lost (sodium perchlorate production) of 2 to 7 grams per ton.
Sodium Perchlorate cell using 'poor' starting solution
The jam jar cell was started up again using a solution containing 95g/l sodium chloride and 496g/l sodium chlorate and around 600ml. This is not a sensible starting solution for Platinum anode perchlorate cells as chloride causes anode erosion. This cell could be called a chlorate cell for the first day or two I suppose. Stirring was not used as the anode was close to the jar bottom and bubbles kept all of the solution moving.
No pH control was attempted. The cell was run at 3 Amps for 3 days and chlorate level was above 15g/l. The pH of the cell was measured at around day 4 and found to be high at around 9. Cell contents were titrated for chlorate level on the seventh day of operation and found to be 5.14g/l Twenty four hours later the chlorate concentration was below 1g/l. The pH at this stage was measure with a high quality probe at a value of 11.5 but you can measure the pH OK using pH paper (only when cells have high pH it would seem, see cell No. 2 below) as there is no bleaching effect as you have in a chlorate cell due to the presence of hypochlorite (bleach!).
Cell contents was evaporated and dried at around 150C for some hours. Total product (presume anhydrous) was 446grams perchlorate. Total Amp-hours into cell was 576Ah AFAICR.

Two more sodium cells
No. 1
Two cells were run using the same starting solution as above, 97g/l chloride and 496g/l chlorate. Cells contained 770ml liquid. Current was 3 Amps. Stirring was used. Chloride was not monitored except to note what the starting concentration was. No pH control was used for the first cell. (Cell No. 2 had acid added)
pH when the chloride became low (cell became a perchlorate cell) was around 9. At 360g/l chlorate, pH was 9.5. Current efficiency was measured at 94% when chlorate concentration was between 360 and 188 grams per litre.
When chlorate concentration was between 188 and 53g/l CE was around 70% with pH around 9.
When chlorate concentration was between 53 and 2.3g/l CE was around 24.5% with pH around 9.6.
Cell was stopped around 14 hours later with chlorate concentration around 0.11g/l.
The cell was run for a total of 7 days.
No. 2
The next cell was the same as above but 12% HCl acid was added to keep pH at around 7 or so for the first few days (chlorate stage). pH control was a bit hap-hazzard. Stirring was used. The pH of the cell was measure at the start and was around 9.8. 2cc acid was added and cell went to pH 7. A few hours later the cell was at pH 9 and it took 19cc of acid to take the pH to 6.8. The next day 7cc of acid was added to keep cell around a pH of 6.8. Next day 7cc acid added. Next day 4cc added very early in the day with no more acid additions needed to keep pH around 6.8. pH drifted down slightly later in the day and chlorate concentration was measured at 400g/l. Later in the day chlorate concentration was at 349g/l. The cell had now ran for 78 hours, 234 Amper-hours had passed in.
When the chlorate concentration was betweeen 349 and 184g/l the CE was 86% with pH around 6.8.
When chlorate concentration was between 184 and 8.8g/l the CE was close to 100% with pH around 4.7. (suspiciously high CE!)
Twelve hours later the chlorate concentration was 0.53g/l with pH now at 1.7. You cannot use pH paper for to measure this low pH for
some reason or other. It gives a false reading of neutral.
This cell had run for 5.5 days to take chlorate concentration to a low level. The last cell ran for six days to take chlorate concentration to a low level.
There was no sign of titanium erosion in the last three cells. All liquor at end of runs was clear.
The Platinium bar now weighed in at 0.91 grams, down from 0.95 three cells/batches ago. That's approx. three batches of 440 grams sodium perchlorate made from a 'not-to-be-recommended' starting solution taken all the way to very low chlorate levels. This equated to 30.3 grams Pt per ton of Na Perk. produced. Ferocious erosion compared to commercial production cells. The starting solution contained chloride. Making perchlorate with chloride in the cell is erosive to platinum according to various sources. Also keeping chlorate concentration high in the cell will give greater CE and probably reduce Pt wear. Using the Pt to reduce the chlorate concentration to very low levels (chlorate scavenging) as opposed to letting the Na Perk. crystallize out of solution (my adding chlorate to cell as the cell progresses) saves lots of work and is easy to do.







corr_pt.jpg - 16kBdisassam.jpg - 82kBkp_on_anode.jpg - 47kBetched_sr.JPG - 54kB

yobbo II - 2-4-2018 at 13:56

Attempted doing the same using tungsten TIG welding electrodes and they work OK.
Can anyone tell me what might be the blue color in the second cell?



Using a Tungsten current runner

Since Titanium is not exactly OTC it was decided to see if Tungsten could be substituted for the Titanium as W is a 'valve' metal similar to Ti, Nb etc. Two TIG welding electrodes of approx. 3.5mm dia. and 150mm long were to hand, the exact type of electrode was unknown (if they contained Thorium, Cerium etc). Thinner electrodes would probably do just as well. The two rods were placed side by side and bound together using plumbers teflon tape. The Pt plate (hammered bar) was slid between the rods and the bottom ends of the rods were then squeezed together tightly with a pair of small vice-grips (locking plyers) and some more teflon tape wound around the bottom ends of the rods and tied securely. A 625 Inconel welding rod was used as the cathode as it was to hand. An ordinary piece of stainless steel would do just as well.
The anode was placed into a cell (similar to the last two cells) of 770ml containing a solution of 95g/l sodium chloride and 496g/l sodium chlorate. Current was three amps. No pH control was attempted. Voltage accross the cell was 6.6V.
After 62 hours of operation the electrode was examined and it was noticed that the tungsten was being etched away. The rods are much thinner as shown in the small picture. There was a yellow compound being deposited on the platinum (WO3, probably). The pH of the solution was around 7.
There was a black deposit on the Inconel cathode above the water line. I have seen this before in perchlorate cells using Nickle cathodes.




Another attempt
Some pure tungsten TIG rods (green tiped) were obtained and a new anode was constructed. A solution of pure sodium chlorate consisting of 230 grams in 770ml, 299g/l, of solution was prepared. There was no chloride present. The anode was similar to the last one except the TIG rods were thinner, at about 2mm dia. Current in the cell was reduced to 2 amps, down from 3. The pH at the start was 5.3. This increased to 7.4 after 22 hours. There was small deposits of yellow stuff appearing on the platinum but it was difficult to see any W corrosion. It was decided to lower the cell pH by adding perchloric acid (around 50%), 3ml taking pH to 1.7 . The yellow deposit disappeared off the platinum in about an hour or so. A few hours later the pH was 0.5 and it stayed at this value for the rest of the time that the cell ran. A brown discolouration appeared in the cell from the start and it was assumed this came from impurities in the starting material.
After seventy one hours of running the chlorate concentration was 4.4g/l. The cell was let run for seven more hours and during this time the solution went completely clear (brown discolouration either deposited on the cell bottom or onto the teflon tape, see picture of electrode). and strangely enough a blue colour appeared. There was also a black deposit on the Inconel cathode.
Cell was run for 6 more hours and closed down. Six hours later the blue colour had gone!
Cell was started up again and the blue colour started to return after 14 hours of running. Five or so hours later it had fully returned (see picture). The colour equivalent was similar to adding about 8 grams copper sulphate pentahydrate to one litre of water, but the actual colour was a more darker, more 'purple' blue. The cell was run for another 21 hours and stopped. One hour later the blue colour was gone, the solution was now clear again. I presume the blue is some Nickle compound???

First Tungsten-runner cell (above) revisited

The saved contents of the cell that I had started with above, (with the thicker Tungsten current runners at 3 Ampers (Tungsten had corroded)), was placed in the cell for to see of there was any W corrosion under a different pH. The pH of the cell contents last time (when W corroded) had been running at around 7.0 . Four cc's of perchloric acid (around 50%) was added to lower pH. A white ppt appeared. pH went to 4.0, 5cc more added (more white ppt) and pH went to 1.6, 10cc more took pH to 1.1 . The solution was now milkey white.
The white ppt may be some Tungsten compound as the cell contents must have some W dissolved (in some form or other) from its time in the first W cell above where corrosion of the W welding electrodes took place. The new pure W (green tiped electrodes) current runners were use here BTW.
Cell contents was 25g/l chloride and 380g/l chlorate. The cell was stirred.
Quite a smell of Chlorine came from the cell, you could actually see the chlorine (I think) sitting on top of the liquid under the lid as you could actually see it blowing away when the lid was lifted and you blew at it! Perhaps there was some ClO2 as well?
Current was 2.05A, Volts was 6.2, Temperature was 28C, cell volume around 790cc.
pH after about 2 hours was 0.6 . pH paper can be used to read pH (Watman Type CF) though I am using a pH meter.
The pH was kept low in a rather hap-hazzard way using 50% (approx.) perchloric acid. The whole problem of trying to keep the pH low is made difficult by the fact that the cell contains chloride. It is ALWAYS best to keep chloride out of perchlorate cells.
Thirty six cc of acid were used in lots of 4cc to 10cc to keep pH around 1.5 for the next 2 days. Sometimes pH managed to get up to 6.0 . There was quite a smell of chlorine after acid was added. The white milkey ppt that was in the cell was inclined to deposit on the Platinum and was brushed off a few times. The black coating that was on the cathode went away. After two days chloride was at around 2g/l and chlorate was at 162g/l. No more acid additions were needed as the cell pH was around 1.3 after the last addition. It slowly rose up to and was about 4.0 at the end of the cell run.
The CE over the next 24 hours was close to 100%.
Some chlorine would have been lost from the system, some added via the perchloric acid. The cell liquid was still not clear, some what milkey appearance. Twenty four hours later the chlorate concentration was 0.4g/l with the liquor getting clearer. There was a white deposit coming onto the cathode. It's as if the ppt that the perchloric acid additions caused had migrated from the solution to the cathode. The anode was clear at this time.
Twenty four hours later the cell was shut down. There was no visible erosion to the W electrodes or that could be detected using a micrometer. The rods are very shiney as if the W oxide coating is transparent. A white deposit appeared on the anode during the last 24 hours, perhaps a W compound from the last run? The Platinum was cleaned and weighed at 0.91g (no loss).



Using a Valve metal works well for platinum (and MMO too). It allows all the surface area of the more expensive and harder-to-get item to be utlized doing something most useful.
It is hard to say if the tungsten erosion above was caused by the presence of chloride, the high pH, the type (purity) of the tungsten electrode or higher current density (50% higher) or a combination of all four. IMO pH has a large effect as there was a yellow deposit appearing on the anode in the second W cell above when the ph was around 7 and this dissolved away without no more erosion to W happened as the cells were run. I am not 100% sure if the yellow deposit was coming from eroded W?
Other valve metals could be used. Tantalum and Hf have better resistance to reverse voltages and therefor will have less risk of getting etched or corroded. Grade 11 Ti (if you can get it) will have greater corrosion resistance. It contains Pd. You will only need a small amount of Valve metal if you keep the anode close to the surface of the cell and use stirring. You could place the cathode at the bottom and the hydrogen would probably keep the cell agigated.
It is impossible to set the pH of a perchlorate cell low if it contains chloride without excessive chlorine gas being generated. As more chloride excapes you need more acid. If using HCl acid this is a vicious circle as you are adding chloride. With perchloric acid it's not much better and you will need to make the acid first. KEEP CHLORIDE OUT OF PERCHLORATE CELLS. Only allow a few grams per litre, or less, to be present.
To quote: "In effect, platinum forms an equilibrium or quasi equilibrium state when evolving chlorine, and a rather different state for oxygen evolution. During co-joint oxygen and chlorine evolution, which occurs in dilute brine, there is interference in the formation of surface layers and this leads to accentuated metal dissolution."
(Platinum Metals Rev., 1998, 42, (l), 27-33)
Something similar is probably going to happen in a perchlorate cell with some chloride in it. The above is in relation to platinum. Perhaps valve metals come under pressure and are more inclined to corrode when used as the anode current runner if chloride is present.
You could use an additive like sodium dichromate as it is supposed to help protect Pt and perhaps the Valve metal current runners too.
Be aware of starting material purity. It was surprising all the colours, ppt's and anode/cathode deposits that occcured when running the cells.
Avoid using nickle cathodes.

Nickle containing cathodes seem to give black corrosion products above and sometimes below the water line. Better to use stainless steel or Titanium.

blue.jpg - 59kBpt_end.jpg - 151kBblack.jpg - 52kB

markx - 3-4-2018 at 00:05

I'm having a suspicion that the blue colour might originate from a form of tungstate (or related compound formed in cell liquor).
I see a very similar discoloration when I electrolyse ammonium molybdate solution at higher current densities to create colored interference layers on stainless steel. The blue appears at the surface of electrodes and dissipates into the solution where it disappears after a few hours. At first I thought this might be caused by dissolution of components from the stainless electrode alloy, but the same appeared also when Ti electrodes were used in the molybdate solution. There is little chance that titanium dissolved under these conditions so I'm inclined to thinking that this is still connected to Mo compounds....as you used a tungsten current lead, there might be a similar situation happening in your cell conditions.

[Edited on 3-4-2018 by markx]

New Anode Material to Try

Simoski - 19-4-2018 at 13:11

Hi

I was watching YouTube a while back and the guy was doing a video on lanthanum. What struck me as interesting was lanthanum hexaboride for use as an anode. It is a ceramic, will not oxidize and is extremely conductive. It should make one of the best perchlorate anodes ever, here an except from Wikipedia...

Lanthanum hexaboride (LaB6, also called lanthanum boride and LaB) is an inorganic chemical, a boride of lanthanum. It is a refractory ceramic material that has a melting point of 2210 °C, and is insoluble in water and hydrochloric acid. It has a low work function and one of the highest electron emissivities known...

Do you think it will produce perchlorate?

yobbo II - 20-4-2018 at 16:30

Two sources state that it is not compatible with strong oxidizing environments/agents. Since an anode IS a strong oxidizing agent I think that would rule it out.


https://www.americanelements.com/lanthanum-hexaboride-12008-...


Lanthanum Hexaboride is highly insoluble in water and converts to the oxide when heated (calcined). Borides are hard, high-melting materials with metal-like conductivity. They are stable to nonoxidizing acids but break down in strong oxidizing agents and strong alkalis. Borides are used in semiconductors, superconductors, diamagnetic, paramagnetic, ferromagnetic, anti-ferromagnetic, turbine blades, and rocket nozzles. Borides have recently been discovered to be superconductive and ultra-incompressible. Lanthanum Boride is generally immediately available in most volumes. High purity, submicron andHigh Purity (99.999%) Lanthanum Boride Sputtering Target nanopowder forms may be considered. American Elements produces to many standard grades when applicable, including Mil Spec (military grade); ACS, Reagent and Technical Grade; Food, Agricultural and Pharmaceutical Grade; Optical Grade, USP and EP/BP (European Pharmacopeia/British Pharmacopeia) and follows applicable ASTM testing standards. Additional technical, research and safety (MSDS) information is available as is a Reference Calculator for converting relevant units of measurement.


https://www.alfa.com/en/catalog/040325/
Notes
Incompatible with strong oxidizing agents.


You can look for a price here but I would not imagine it will be cheaper that Pt.

https://www.lesker.com/newweb/deposition_materials/depositio...

and here

http://www.samaterials.com/lab6/1413-lanthanum-hexaboride-la...



More info</a></li></ul><div class="sheets align_justify" id="more_info_sheets"><div class="rte" id="idTab1"><p>Lanthanum Hexaboride (LaB6, also called lanthanum boride and LaB6) is an inorganic chemical. With melting point 2528 K, LaB6 is a refractory ceramic material, which is insoluble in water or hydrochloric acid and very stable in vacuum. Stoichiometric samples are colored intense purple-violet, while boron-rich ones (above LaB6.07) are blue. Ion bombardment changes its color from purple to emerald green. LaB6 is a superconductor with a relatively low transition temperature of 0.45 K.</p><p><strong>Specification:</strong></p><table border="0" cellpadding="0" cellspacing="0" style="width: 684px;"><tbody><tr><td width="160"><p align="left"><strong>Product</strong></p></td><td width="168"><p align="left">Lanthanum Hexaboride</p></td><td width="191"><p align="left"><strong>Structure</strong></p></td><td width="165"><p align="left">Polycrystalline</p></td></tr><tr><td width="160"><p align="left"><strong>Symbol</strong></p></td><td width="168"><p align="left">LaB6</p></td><td width="191"><p align="left"><strong>Thermal Conductive </strong></p></td><td width="165"><p align="left">47 W/mK (20℃)</p></td></tr><tr><td width="160"><p align="left"><strong>Cas No.</strong></p></td><td width="168"><p align="left">12008-21-8</p></td><td width="191"><p align="left"><strong>Thermal Expansion </strong>&nbsp;<strong></strong></p></td><td width="165"><p align="left">6.2 10<sup>-6</sup>K<sup>-1</sup> (20-900℃)</p></td></tr><tr><td width="160"><p align="left"><strong>Atomic Mass</strong></p></td><td width="168"><p align="left">203.78 g/mol</p></td><td width="191"><p align="left"><strong>Electrical Resistance</strong>&nbsp;<strong></strong></p></td><td width="165"><p align="left">ca.15 μΩ cm (20℃)</p></td></tr><tr><td width="160"><p align="left"><strong>Density </strong></p></td><td width="168"><p align="left">4.72 g/cm<sup>3</sup></p></td><td width="191"><p align="left"><strong>Electrical Conductive</strong></p></td><td width="165"><p align="left">6.65x10<sup>4 &nbsp;</sup>S/cm (20℃)</p></td></tr><tr><td width="160"><p align="left"><strong>Melting Point</strong></p></td><td width="168"><p align="left">2528 K</p></td><td width="191"><p align="left"><strong>Currenty Denstiy </strong></p></td><td width="165"><p align="left">150 A/cm2 (1950℃)</p></td></tr><tr><td width="160"><p align="left"><strong>Hardness</strong></p></td><td width="168"><p align="left">87.5 RA</p></td><td width="191"><p align="left"><strong>Electron Emissivity</strong></p></td><td width="165"><p align="left">2.6 eV</p></td></tr><tr><td width="160"><p align="left"><strong>Flexure Strength (σ)</strong></p></td><td width="168"><p align="left">200 Mpa</p></td><td width="191"><p align="left"><strong>Fracture toughness (Kic)</strong></p></td><td width="165"><p align="left">3.0 MN/m<sup>3</sup>/2</p></td></tr></tbody></table>

Yob

yobbo II

Sulaiman - 21-4-2018 at 03:54

It seems to me that you have been operating your cells with a significant surface area of titanium exposed to the electrolyte(s),
comparable to the platinum surface area,
yet there seems to have been only moderate corrosion of the titanium.

If this observation is correct, why not just use titanium for the anode ?

(I've not made a chlorate or perchlorate cell yet, but it's on the list.

markx - 21-4-2018 at 09:34

Quote: Originally posted by Sulaiman  
It seems to me that you have been operating your cells with a significant surface area of titanium exposed to the electrolyte(s),
comparable to the platinum surface area,
yet there seems to have been only moderate corrosion of the titanium.

If this observation is correct, why not just use titanium for the anode ?

(I've not made a chlorate or perchlorate cell yet, but it's on the list.


Titanium is what is called a "valve metal"...it passivates under anodic conditions and becomes covered by nonconductive oxide layer, hence it also resists corrosion very successfully. Ti can not be used as the anode on its own, only as a substrate material and current lead.

yobbo II - 22-4-2018 at 12:24


The titanium only corroded when used in a potassium perchlorate cell and in a strontium perchlorate cell. In the case of the K Perk cell, the ionic strength (total dissolved solids) of a K perchlorate cell gets low and this (IMO) stresses the titanium oxide and you get corrosion.
In the case of the Sr Perk cell I was using a piece of 'dodgy' titanium that had been subject to high heat some time in the past. This may have effected its ability to withstand corrosion (poorer oxide coat) or perhaps Sr Perk cells have a greater negative effect on the protective Ti oxide layer.

The whole point of using a valve metal as the current runner to transport current to the platinum anode (it's the Pt that is the anode, not the Titanium) is that you can purchase a small piece of platinum and use ALL of it to do useful work.
Platinum wire is inclined to be very expensive and using the (usually fairly thin) wire to transport all the current to the Pt below the surface of the liquid leaves the wire very hot (the piece above the liquid surface) and liable to corrode or even melt and break off.

stamasd - 16-6-2018 at 04:00

Let me start documenting here my experiments in making potassium perchlorate from potassium chlorate using MnO2 anodes.

I made my anode from a 5*10cm plate of Ti (0.5mm thick) cut from a larger sheet of 10*10cm; the other part of the sheet was cut in strips and the strips used to make the cathode for the cell as well as the connecting terminal for the anode. One strip of 1.5*10cm became the cathode, another strip of the same size was spot-welded to the anode after coating using a spot welder I have (used normally for thin nickel strips, but on the highest energy setting works well for thin titanium too).

For the coating I used thermal decomposition of cobalt nitrate and manganese nitrate. Both reagents were made from pottery-grade cobalt and manganese carbonates. The cobalt nitrate appeared pure enough, the manganese one had iron contamination and was further purified using the method from NurdRage's old video (partial conversion to hydroxides, overnight oxidation etc). I did 4 applications of cobalt to each side of the anode (after thoroughly degreasing, mechanical scouring and chemical etching in dilute nitric acid) followed by 6 applications of manganese to each side. Thorough rubbing with a wet paper towel after each application to remove any loose adhering oxides. The anode was then dried and then again rubbed thoroughly with moist towel until no more material rubbed off of it mechanically.

A preconditioning of the anode was done by running in a 10% NaCl solution at approx 3A for a few hours. The solution turned a light pink, but no change in the physical appearance of the anode was noted, and no particles of MnO2 appeared to shear off it. The electrolyte remained clear.

For the actual experiment I used KClO3 previously produced in a Ti/MMO cell and purified. 122g KClO3 (roughly 1 mol) was dissolved in warm distilled water to a final volume of 950ml; that was enough to keep all KClO3 dissolved at 40 degrees C. I ran the cell at an average of 3A (the current varied during the electrolysis because of evaporation and relative positioning of the electrodes, but the 3A is IMHO a good estimate of the average; min current was 2.3A, max current 4.4A). No pH control was done. A total of 200ml water was added during the run. The time was 33.5h, and final volume of the electrolyte was just above 800ml. A layer of acicular crystals was at the bottom of the cell (about 2cm thick in a 1L beaker), and a substantial amount of the same crystals was floating at the top but were easily dislodged by shaking and sank to the bottom. The crystals were mixed with a small amount of dirty brown powdery residue, likely MnO2. During the electrolysis, the color of the electrolyte changed from colorless to a light pink immediately after the start, the pink color intensified in the first 3-4h to a maximum pink-purple then began to fade very slowly to a final light pink. Too bad I didn't think of taking pictures of each stage.

Total current used was approx 100Ah; theoretical needed at 100% current efficiency for 1mol KClO3->KClO4 conversion is 2*26.8=53.6AH so I used 1.86 times that. This was done deliberately as I expect the current efficiency to be 50-60%.

The anode appearance at the end was pretty much unchanged from the start. The cathode was covered with a dark brown sludge, most of which was easily rubbed off with a paper towel, and anything remaining after that came off with light scouring with a copper sponge.

My purification strategy for the product is as follows:

1. removal of the dark brown precipitate: I avoided the need for a hot filtration (which has been problematic in the past) by heating the electrolyte until all crystals dissolved, allowed it to stand hot in the beaker and doing repeated decanting with short blasts in the microwave in between to keep the solution hot. I achieved substantial removal of the precipitate this way. The solution was allowed to cool to RT and a layer of acicular crystals formed again at the bottom. Will attach picture below.

2. From this point on, I will do the following (still to come). I made the assumption that no more than 50% of the chlorate is still in the electrolyte, unconverted. That's 61g from the 122g initially added. The minimum amount of solution that will keep that dissolved at 25C is 750ml, convenient since I have 800ml now. I hope that the crystals at the bottom are mostly KClO4, with only trace KClO3. After allowed to mature for 1 day, the crystals will be filtered out, quickly rinsed with water and dried. I will not further process the remaining electrolyte now, but will keep it to use as base for further electrolysis.

3. Will need to test the crystals for chlorate. I don't have now any of the usual reagents for the tests, nor to funds to purchase them. When I can I will get some indigo carmine and use that method. For removal of chlorate I plan on using the metabisulfite method, which I have plenty of (I'm a homebrewer and use that for wine). After further recrystallization and testing for chlorate I will calculate the yield. That will take a while though as I have to be at work all of this weekend and all of next week.

So far so good, the anode has degraded much less than I expected, and the contamination by MnO2 and KMnO4 also less problematic than I thought it would be. But again, I will have to get a yield calculated.

Below pictures of the anode and cathode after the run, and of the crystals formed after removal of the precipitate.

20180616_060428.jpg - 1.5MB 20180616_060448.jpg - 1.5MB 20180616_055754.jpg - 1.5MB


[Edited on 16-6-2018 by stamasd]

stamasd - 17-6-2018 at 15:17

Update: a rough weighing of the now almost completely dry crystals showed 57g. Even accounting for the estimated 12g or so of perchlorate still in the electrolyte, that still makes a terrible yield.

stamasd - 18-6-2018 at 06:48

So I'm sitting here during a break at work and planning the next experiment. Here's a quick overview, and a question.

1. reuse the 800ml electrolyte from previous experiment to benefit from the residual chlorate and perchlorate already in it.
2. add another mol of KClO3, heat until it dissolves, adjust volume to 1000ml, let cool, pour off anything that crystallizes.
3. Make a better cathode; most of the problems I've had in the previous run were due to the short cathode. Will spot-weld another strip of titanium to the existing one.
4. ?? (see below)
5. run at 3A, 48h which should be 2.68x the current needed at a theoretical 100%CE.
6. see what yields.

The ?? at #4 above pertains to using additives in the cell. Candidates would be NaF or K2Cr2O7. I've read that NaF is not recommended for any cells using Ti-substrate anodes. This leaves the dichromate, but either my search-fu isn't working or nobody has used it in a cell with MnO2 anodes.

Does anyone have data on K2Cr2O7 used in Ti/MnO2 cells?


(edit) I have started the 2nd experiment thus:


800 ml electrolyte from previous expt; added 123g KClO3 and heated, added water to 1100ml (not everything dissolved, there are about 10-20g crystals on bottom)
Added 2.6g Na2Cr2O7 (turns out I don't have K2Cr2O7). Changed cathode to 165*50mm Ti mesh (it's about 2/3 immersed in electrolyte). Put magnetic bar in and started stirring on med.
Current limited to 5A. Voltage is 5.7V.
Start time is 6/18/18 at 7PM
24h=24*5=120Ah; that is 120/53.6=2.23x the theoretical.
Will stop after 24h and see what results.
Will be hard to appreciate any permanganate formation because the color is bright yellow from the dichromate.

New cathode picture below.

20180618_185550.jpg - 1.4MB

Also picture of the cell at 12h. The stirbar was overwhelmed by the crystals at the bottom and I couldn't get the stirring restarted.



20180619_061235.jpg - 2.3MB

[Edited on 19-6-2018 by stamasd]

stamasd - 19-6-2018 at 15:58

After 24h and 120Ah the color turned to a muddy yellow-green. I guess a lot of Cr3+ was formed. Picture below. The cathode got a grey tint that doesn't come off with scrubbing. The anode appears mostly OK even though the layer of oxides appears somehow more translucent than before. Still functions well as an anode. The current just before shutdown was 5A, voltage had risen to 6.35V.

20180619_194947.jpg - 1.5MB 20180619_190529.jpg - 2.4MB 20180619_190515.jpg - 2.4MB

I've put the beaker in a fridge; I won't have time to process what's inside for several days.

[Edited on 20-6-2018 by stamasd]

NedsHead - 1-7-2018 at 00:52

I found some Pt plated titanium for sale on eBay from a German seller and thought I would post it here for anyone that might be interested, I can't vouch for them myself but they do have a good seller rating and have some other anodes at a good price.

https://www.ebay.com.au/itm/Platin-Elektrode-Platiniertes-Ti...

hissingnoise - 1-7-2018 at 02:23

70mm of plated anode seems a bit, er, on the short side...?


NedsHead - 1-7-2018 at 03:06

it's 80mm, and it might be useful to someone

JScott - 30-8-2018 at 14:01

I promise, better pictures next time. I didn't think anyone would see them when I took em ;-)

I followed NerdRage's video and his advise regarding platinum anodes. I had initially run this with a dc/dc converter that offered voltage and current control but found it was horribly inefficient. I will rethink the converter for the next run.

Straight from the five volt pin of my little breakout box it used no more than 2-3 amps over the few days I ran it. I am away from notes tonight, but tomorrow will post more info regarding yield over time.

My wife found these containers at Wally world. I've seen others online have adopted them for cells also.

The anodes (platinum plated titanium) we purchased directly from the link NerdRage provided. He (TMC if I remember correctly) is still selling them on Ebay and Amazon. It held up very well over the seven days I ran this cell. My cathode was a titanium strip purchased from the same seller. It also held up very well. Other than discoloration, there is little evidence they have even been used. Some slight pitting where both entered the solution.

The digest from my notes I promised yesterday:

The cell held 1300ml. The solution was distilled water saturated with chloride (1.16 grams/ml). I found it difficult at the labs ambient temperature (22.5C in the hood on this day) to keep this much chloride in solution. In my first run a good deal of chloride fell out of solution, perhaps as the lab cooled at night. I mistaken believed at first I had a banner overnight success only realize it wasn't cholrate collected at the bottom of the cell.

The cell seemed, through evaporation, to loose about a ml a day. I topped it off each day (with saturated solution) and checked it's saturation levels. It seems by adding that milliliter of saturated solution each day, I kept the cell well feed. The temperature never exceeded 52.7C. The voltage was between 3.2 and 3.5 volts and I the cell drew 2-3 amps. I did not account for cell Ph, but will remember to do that next run.

I collected 192.7 grams in seven days.



cell_1.jpg - 73kB

[Edited on 8-30-2018 by JScott]

[Edited on 8-31-2018 by JScott]

Simoski - 27-11-2018 at 11:23

Here are some thoughts on anodes, I have been making chlorates for a few years now and have some insight I would like to share.

Initially I used carbon gouging rods with 316l ss cathodes, I would build arrays of like 20 of them ( 20 rods ) and ran the electricity in and out from the top. This is, as far as I have seen, a very common practice and quite logical. However what I found was that the carbon rods corroded right at the top of the cell. This too makes sense because the electrons enter the cell at the cathodes and take the path of least resistance through the electrolyte to the anode and exit. The current density at the top of the cell was very high and was causing the corrosion . Later versions ran electrons in at the bottom of the cell and out at the top and this extended the life of the electrode assembly.

Later I swapped to mmo and still use this principle today, I run the current down to the bottom of the cell via the cathode but cover the stainless steel in heatshrink until the bottom, where I allow the steel to contact the electrolyte. This will reduce localised high current densities at the top of the cell and extend the life of you electrode assemblies.

This is another tip I have for anyone out there creating chlorate cells, not so much a thought on anodes but an insulating chemically stable thermoplastic that you can use to cover copper interfaces and run current down into a very corrosive electrolyte ... heatshrink!

[Edited on 27-11-2018 by Simoski]

[Edited on 27-11-2018 by Simoski]

pinko - 24-12-2018 at 10:43

Here is a description of my experience making Pt anodes. It took some time mastering the process but the results are excellent.

http://www.blog.exrockets.com/blog/making-platinum-foil-and-...

Welding Pt-Pt and Ti-Pt for the home scientists is not difficult. Pt-Pt welds are done with graphite and Ti-Pt welds using DIY spot welder.

http://www.blog.exrockets.com/blog/diy-spot-welder/

Hope the information will be helpful.

morganbw - 24-12-2018 at 15:28

Quote: Originally posted by pinko  
Here is a description of my experience making Pt anodes. It took some time mastering the process but the results are excellent.

http://www.blog.exrockets.com/blog/making-platinum-foil-and-...

Welding Pt-Pt and Ti-Pt for the home scientists is not difficult. Pt-Pt welds are done with graphite and Ti-Pt welds using DIY spot welder.

http://www.blog.exrockets.com/blog/diy-spot-welder/

Hope the information will be helpful.


Interesting. Thank you.

[Edited on 12/24/2018 by morganbw]

yobbo II - 25-12-2018 at 18:12

How thick is the Pt foil that you welded to the titanium?

Cheers,
Yob

pinko - 25-12-2018 at 18:39

Quote: Originally posted by yobbo II  
How thick is the Pt foil that you welded to the titanium?

Cheers,
Yob


I've made electrodes by spot-welding between half a thou (~13 microns) and 3 thous (~75 microns) Pt foil on a 0.5mm and 1mm thick Ti strips.

What should be tested in advance is the welding current. Too much blows a hole in the Ti strip (as shown in one of my pictures), too little doesn't melt and fuse the Pt foil.

I also found out that double (with preheating) spot-welding pulses work better for proper Pt-Ti fusion.

[Edited on 26-12-2018 by pinko]

wg48 - 25-12-2018 at 20:18

Quote: Originally posted by pinko  
Quote: Originally posted by yobbo II  
How thick is the Pt foil that you welded to the titanium?

Cheers,
Yob


I've made electrodes by spot-welding between half a thou (~13 microns) and 3 thous (~75 microns) Pt foil on a 0.5mm and 1mm thick Ti strips.

What should be tested in advance is the welding current. Too much blows a hole in the Ti strip (as shown in one of my pictures), too little doesn't melt and fuse the Pt foil.

I also found out that double (with preheating) spot-welding pulses work better for proper Pt-Ti fusion.

[Edited on 26-12-2018 by pinko]


pinko@ That rolling your own Pt foil from a small ingot was heroic work and you made the rolling machine yourself too. *** (three gold stars)

yobbo II - 25-12-2018 at 21:00


The rolling machine in a work of art, never mind the science!

A comment from your page by Pinko:

"If there is no real fusion between the metals [the Pt and Ti], unlike the spot-welding where both are alloyed in the weld, eventually some oxygen will creep to the Ti surface and it will get oxidized which is going to act as an insulator and the electrode will fail."

This is actually not true. The current will flow across a joint that is made with titanium and Pt tightly pinched together.
I believe the non conducting Ti oxide that is surly there? is made conductive by the presence of small amounts of hydrogen.
There is a paper somewhere or other explaining this phenomenon.

Yob

pinko - 25-12-2018 at 21:35

Quote: Originally posted by yobbo II  

I believe the non conducting Ti oxide that is surly there? is made conductive by the presence of small amounts of hydrogen.
There is a paper somewhere or other explaining this phenomenon.
Yob


Very interesting, thank you for sharing this information. Now when you mentioned that, I remember that TiO2 is behaving as a semiconductor (~3.5 eV band-gap *Google).

Could it be that the Hydrogen diffuses in TiO2 thus lowering the band-gap, interesting if there will be voltage drop in the junction? Also any idea what would be the resistance in such joints?


[Edited on 26-12-2018 by pinko]

yobbo II - 28-12-2018 at 06:13



This paper talks about the it, click the pdf link.
https://www.hindawi.com/journals/apec/1974/398185/abs/

There may be others.

When you use Ti as a cathode the current flows OK. The Ti Oxide has not gone away but the H generated at the cathode makes it conductive
unlike the anode side where there is no H.
When you pinch the Pt between the Ti plates there must be some small amounts of H 'in there somewhere' to make the oxide conductive.
That's my limited understanding of it.

Yob

Simoski - 2-1-2019 at 08:19

Quote: Originally posted by pinko  
Here is a description of my experience making Pt anodes. It took some time mastering the process but the results are excellent.

http://www.blog.exrockets.com/blog/making-platinum-foil-and-...

Welding Pt-Pt and Ti-Pt for the home scientists is not difficult. Pt-Pt welds are done with graphite and Ti-Pt welds using DIY spot welder.

http://www.blog.exrockets.com/blog/diy-spot-welder/

Hope the information will be helpful.


Cool Pinko, thanks

Good chlorate anode materials?

Simoski - 2-1-2019 at 08:30

Yobbo II or anyone else do you think the following materials will make a good anode ...

lanthinum calcium manganate

or

lanthinum strontium manganate?

( the latter being used as a cathode in solid oxide fuel cells (SOFC) )

Simoski - 16-2-2019 at 14:45

Electrolysing scrap steel to make iron oxide has taught me this about steel anodes, cast iron disk brakes last much longer.

Put another way cast iron is far more inert than mild steel.

Gearhead_Shem_Tov - 16-2-2019 at 16:20

Quote: Originally posted by Simoski  
Electrolysing scrap steel to make iron oxide has taught me this about steel anodes, cast iron disk brakes last much longer.

Put another way cast iron is far more inert than mild steel.


Would that be because of cast iron's much higher carbon content do you suppose? Cast iron has about ten times the carbon content as mild steel.

-Bobby

Simoski - 30-3-2019 at 11:14

Quote: Originally posted by Gearhead_Shem_Tov  
Quote: Originally posted by Simoski  
Electrolysing scrap steel to make iron oxide has taught me this about steel anodes, cast iron disk brakes last much longer.

Put another way cast iron is far more inert than mild steel.


Would that be because of cast iron's much higher carbon content do you suppose? Cast iron has about ten times the carbon content as mild steel.

-Bobby


Bobby I am not sure exactly it just seems to me that it is more dense, more solid.

Simoski - 26-4-2019 at 21:47

There is a channel on YouTube where a guy squashes all sorts of things in his industrial press. This got me thinking... what if we took MnO2 powder and squashed it under immense pressure ( like 50 tons / square inch ) would it fuse into a solid piece that we could use as an anode? Could we do the same to PbO2 powder?

[Edited on 27-4-2019 by Simoski]

Simoski - 2-7-2019 at 22:44

What about Zirconium as an inert anode for chlorate / perchlorate production, will it passivate like titanium?

Let me answer my own question:

Zirconium is a member of the family of 'valve metals' (an archaic term), which form a strongly passivating oxide film when exposed to air and/or water. From here:

Owing to their low electrochemical potential the group IVB and VB valve metals Ti, Zr, Hf, V, Nb and Ta readily react with water or oxygen to form a dense, protecting passive layer.




[Edited on 3-7-2019 by Simoski]

Simoski - 29-7-2019 at 07:39

Quote: Originally posted by Simoski  

Zirconium is a member of the family of 'valve metals' (an archaic term), which form a strongly passivating oxide film when exposed to air and/or water. From here:

Owing to their low electrochemical potential the group IVB and VB valve metals Ti, Zr, Hf, V, Nb and Ta readily react with water or oxygen to form a dense, protecting passive layer.
[Edited on 3-7-2019 by Simoski]


I now wonder if Zr, Hf, V, Nb and Ta would make as a good a cathode in a chlorate cell as Ti? I wonder because Ti is an excellent Chorate Cell cathode.
Ti passivates as an anode but makes a very stable cathode.




[Edited on 29-7-2019 by Simoski]

Electron Well

Simoski - 29-7-2019 at 07:54

Since it seems to me to carry these thoughts forward, let me continue by saying thank you to all those before and after.

BUT HEAR ME NOW:

THE ANODE IS AN ELECTRON WELL... a sink for electrons, down you go electrons, down you go!

One can imagine being miniturised to the point where an electron was the size of your head, then standing on or against an anode immersed in an electrolytic cell. Looking out through the electrolyte one would see these football sized "electrons" moving through the electrolyte and then sinking into the anode, gone.... ??? no but the appearance of gone, rather absorbed for further transmission.

The cathode is therefore an electron font or spring or source.


[Edited on 29-7-2019 by Simoski]

mysteriusbhoice - 26-4-2020 at 01:38

conductive epoxy will fail because the epoxy actually degrades in solution as the cell runs so you need to impregnate a plastic that wont degrade one such would be whatever your cell is made of HDPE so that would be a good choice to mix conductive graphite mud/powder into

Tamerlane - 31-7-2020 at 14:09

Here's an interesting article which uses a particle bed anode of lead dioxide to make perchlorate, with an MMO current feed. An amateur-friendly idea could be to employ a trough or pan with an MMO mesh placed at the bottom and a PbO2 "sand" pressed in above it.

mysteriusbhoice - 7-12-2020 at 22:40

Quote: Originally posted by Tamerlane  
Here's an interesting article which uses a particle bed anode of lead dioxide to make perchlorate, with an MMO current feed. An amateur-friendly idea could be to employ a trough or pan with an MMO mesh placed at the bottom and a PbO2 "sand" pressed in above it.


the thing with this cell is that you need a good diaphragm pump to actually flow liquid through the cell also even if its just standing you would need tons of PbO2 powder and at that point its way better to use epoxy,acrylic,PVC and PbO2 composite electrodes since those do work just use recyled Ti as substrate and not graphite!!

flaminglasrswrd - 7-3-2021 at 20:03

Quote: Originally posted by mysteriusbhoice  
Quote: Originally posted by Tamerlane  
Here's an interesting article which uses a particle bed anode of lead dioxide to make perchlorate, with an MMO current feed. An amateur-friendly idea could be to employ a trough or pan with an MMO mesh placed at the bottom and a PbO2 "sand" pressed in above it.


the thing with this cell is that you need a good diaphragm pump to actually flow liquid through the cell also even if its just standing you would need tons of PbO2 powder and at that point its way better to use epoxy,acrylic,PVC and PbO2 composite electrodes since those do work just use recyled Ti as substrate and not graphite!!


Some quick calculations:
The original article used 5kg of PbO2 in a 5cm layer in a 20cm ID HDPE pipe and 80 amps. That's a current density of about 250ma/cm^2 with PbO2 packed at a density of about 3 g/cm^3. They achieved 80-90% CE (20-25% higher than conventional parallel plate systems).

At 1/16th scale: 2" PVC pipe packed to 5cm with about 300g of lead dioxide would operate with about 5A. PVC is resistant to chlorinating liquids. Standard plumbing fittings make assembly easy. Making PbO2 electrochemically from dirt cheap lead metal is also easy if you are actually trying to get particles.

Flow rates would need some adjusting because the scaling probably isn't linear, but something around 50 ml/min is reasonable. That's achievable with air-lift pumps which are completely chemical resistant. For better efficiency, a pneumatic ejector pump can be used.

Honestly, I think this is a great option that I'm going to pursue at some point.

STUDY ON AIRLIFT PUMP AS A PUMPING AND AERATION SYSTEM IN AQUACULTURE

yobbo II - 25-3-2021 at 12:12

The Lead Dioxide bed is somewhat similar an Anode make my G. Pinkston long time ago.
http://www.chlorates.exrockets.com/leaddiox/ldslda.html


Changing the subject slightly to DSA, there is an attached article.

Yob

Attachment: denki_anodes.pdf (811kB)
This file has been downloaded 414 times


yobbo II - 11-11-2021 at 15:10

Some more reading.

Yob

Attachment: denki_anodes.pdf (811kB)
This file has been downloaded 331 times


mysteriusbhoice - 23-1-2022 at 23:54

some stainless steel coated with PbO2-CeO2 nanocomposite layer will soon test if its viable.

PbO2 on SS.jpg - 22kB

B(a)P - 24-1-2022 at 01:02

Quote: Originally posted by mysteriusbhoice  
some stainless steel coated with PbO2-CeO2 nanocomposite layer will soon test if its viable.



Looking forward to hearing how it goes. Why did you go with a metal ruler as a substrate?

mysteriusbhoice - 24-1-2022 at 08:15

Quote: Originally posted by B(a)P  


Looking forward to hearing how it goes. Why did you go with a metal ruler as a substrate?


Because its what I had on hand and overall its a really cheap 10 cent steel stick so why not and tho ive established Ti as a viable substrate I really wanna move onto other substrates which are more accesible like copper for instance.

The main issue I have with stainless steel is adhesion of the PbO2 coating it just doesnt seem to want to stick to stainless steel maybe because its a flat plate.

It also has issues sticking to Ti plates and seems to only want to stick to mesh type electrodes maybe due to having more surface area to latch onto.

mysteriusbhoice - 6-11-2022 at 06:47

so then I used cerium tin oxide as a coating on HF etched Ti shown in this video series.

part1: substrate preperation.
https://www.youtube.com/watch?v=HG6ZLxmd4MU

part2: painting DTO and baking using blow torch.
https://www.youtube.com/watch?v=cXH1qT3k-9k

Part3: coating of overcoat of PbO2(optional) increases lifetime and improves conductivity TBA

mysteriusbhoice - 6-11-2022 at 06:47

so then I used cerium tin oxide as a coating on HF etched Ti shown in this video series.

part1: substrate preperation.
https://www.youtube.com/watch?v=HG6ZLxmd4MU

part2: painting DTO and baking using blow torch.
https://www.youtube.com/watch?v=cXH1qT3k-9k

Part3: coating of overcoat of PbO2(optional) increases lifetime and improves conductivity TBA

Sir_Gawain - 28-3-2023 at 18:42

For some reason, Ebay has tons of platinized titanium anodes right now like this. I doubt they are real because of the price. Has anyone tested any like this? They ship from the US and have free returns, so I might buy one and test it.

mysteriusbhoice - 28-3-2023 at 22:52

Quote: Originally posted by Sir_Gawain  
For some reason, Ebay has tons of platinized titanium anodes right now like this. I doubt they are real because of the price. Has anyone tested any like this? They ship from the US and have free returns, so I might buy one and test it.


Anyhing that looks shiny like that is likely to be a fake because real baked on or plated on Pt shows a rough surface finish because its plated or baked onto an etched Ti substrate and will never be shiny and smooth in finish.

Sir_Gawain - 29-3-2023 at 16:42

Are you sure plated platinum isn't shiny? This guy made some and they look pretty shiny;https://www.youtube.com/watch?v=TSzgvP0OW-Q. (I still think the Ebay ones are fake though.)

Rainwater - 29-3-2023 at 16:51

4 different sellers, all the same thing.
Steel mesh with nickel plating.
None lasted more than a few seconds in a chlorate cell before discoloring.
All where magnetic
None decomposed h2o2.
Full refunds on all of them
Exact same picture and description.

[Edited on 30-3-2023 by Rainwater]

mysteriusbhoice - 29-3-2023 at 17:03

Quote: Originally posted by Rainwater  
4 different sellers, all the same thing.
Steel mesh with nickel plating.
None lasted more than a few seconds in a chlorate cell before discoloring.
All where magnetic
None decomposed h2o2.
Full refunds on all of them
Exact same picture and description.

[Edited on 30-3-2023 by Rainwater]


Honestly I have a full guide on making capable electrodes from scratch using only Ti mesh by first plating DTO using an aqueous prep without the need for dry Cl2. and then plating PbO2 over tht and doing pH non controlled at 11 and temp below 40C so the hypochlorite stays high so that the electrode doesnt errode.

mysteriusbhoice - 30-3-2023 at 14:54

Quote: Originally posted by Sir_Gawain  
Are you sure plated platinum isn't shiny? This guy made some and they look pretty shiny;https://www.youtube.com/watch?v=TSzgvP0OW-Q. (I still think the Ebay ones are fake though.)


yea and they didnt work so he bought commercial ones.
He didnt make them properly and I know that because I spent years trying to make electrodes and now know how proper anodes behave and should look like. If you want real electrodes go to this link
https://www.aliexpress.com/item/1005002682948243.html
Pt is overrated anyway and Ir-Ta MMO can do everything Pt can do except perchlorates but this is fixed by doping with Pb.
also below picture of genuine Pt electrode which is shown as having a clear divide between the Ti and Pt layer.

pt1.png - 63kB

Sir_Gawain - 1-4-2023 at 15:42

Ir-Ta can do chlorate to perchlorate?

mysteriusbhoice - 1-4-2023 at 22:43

Quote: Originally posted by Sir_Gawain  
Ir-Ta can do chlorate to perchlorate?


if doped with PbO2 yes otherwise its only for chlorate and wont go further.
use alkaline plumbite bath for doping 40g/l PbO and 160g/l NaOH 2 hours at 2.3ma/cm^2.

chemistry4frog - 21-8-2023 at 15:52

can bismuth be used as an anode for chloro alkali. i was thinking a bismuth ignot as an anode for chloro aklali ??
 Pages:  1  ..  10    12