Sciencemadness Discussion Board
Not logged in [Login ]
Go To Bottom

Printable Version  
Author: Subject: new rotovap or improvise
andre178
Hazard to Self
**




Posts: 61
Registered: 11-12-2009
Location: San Diego, CA
Member Is Offline

Mood: pacified

[*] posted on 7-2-2011 at 11:43
new rotovap or improvise


I'm curious, after a bit of searching on google, and some major manufacturers, I found that new rotovaps are hard to purchase, if not impossible. Is it a technology that can be improvised or gone around using a condenser and stirring rod?

The old one I'm using right now is slowly giving out, it leaks everywhere, and I'm not sure if I should spend time replacing it or just making a tilted system with a condenser upright improvisation for distilling solvents.
View user's profile View All Posts By User
Magic Muzzlet
Hazard to Others
***




Posts: 146
Registered: 22-7-2010
Member Is Offline

Mood: No Mood

[*] posted on 7-2-2011 at 12:26


Well you could get one off eBay or labx, or you could just use a rapidly spinning stir bar and remove your solvent in a distillation setup. This is what I do, it takes just a bit longer but it's all the same. No way in hell I'm paying thousands or even hundreds for rotovap when I can do the same with equipment I already have. ;)
View user's profile View All Posts By User
not_important
International Hazard
*****




Posts: 3873
Registered: 21-7-2006
Member Is Offline

Mood: No Mood

[*] posted on 7-2-2011 at 15:27


Early 20th century tech was a pair of flasks, connected using a pair of stoppers and a wide glass tube so as to resemble a cartoon barbell. Laid on their sides, one flask in a hot water bath and the other in cold/ice water, possibly with a stream of cold water running over the cold flask. The flasks were rotated around the axis formed by the tube, in some cases using a simple water turbine on the tube near the condensing flask and turned using a 2nd stream of water. The fancy setups had a small tube offset in one the the stoppers, a rubber tube attached to that and through which the flasks were evacuated and then the rubber tube clamped closed.

So, early versions of rotovaps, not as effective with the really low boiling solvents.

View user's profile View All Posts By User
Lambda-Eyde
National Hazard
****




Posts: 857
Registered: 20-11-2008
Location: Norway
Member Is Offline

Mood: Cleaved

[*] posted on 7-2-2011 at 15:38


Having a rotavap in the lab makes life easier for you. Not having to deal with clips, clamps and tubes every time you want to remove a solvent is preferable - that's why it's so nice to have a rotavap ready for work at all times! If I had one with a few problems, I'd try to find replacement parts for it. They can be quite expensive, but MUCH less expensive than buying a new rotavap.
View user's profile View All Posts By User
peach
Bon Vivant
*****




Posts: 1428
Registered: 14-11-2008
Member Is Offline

Mood: No Mood

[*] posted on 8-2-2011 at 00:41


There is a PDF floating around google somewhere written by a guy at university (a lecturer I think) who was complaining about the cost of rotovaps.

He'd made one using a ball and socket joint he'd turned on a lathe from PTFE, sealed with vacuum grease.

If you've got a lathe to hand and are good with your hands, you could try that. If not, I wouldn't try randomly connecting bits of glass and motors together.

One possibility to produce the vacuum tight, but rotating, socket may be to use an overhead stir gland, as these are designed to maintain a vacuum whilst also spinning - with one section staying still.

A side effect of using a stir gland is that they are usually glass. Balancing a heavy flask full of solvent on one end and spinning it, you would need to be sure of what you were doing or I can envisage accidents fast approaching.

As some of the others have already pointed out, you can remove solvent very easily using standard glassware and a vacuum. Even without the vacuum, most of the volatiles will come off quickly.

The benefit of the rotovap is simplicity. Meaning, if you have to be incredibly careful with your DIY setup, you may be better just using a normal still setup.

I am also a little upset by rotovaps, as they seem to be promoting people to just ignore what is going on. I have seen numerous complaints from people about solvent getting through the traps. This is always a side effect of them running them from high vacuum sources.

I suspect the older generations, more used to having to set up a still and put a bit more care into it, will appreciate why the solvent may be getting through the trap.

Worse, the solutions I see to the problem involving spending more money as opposed to less.

Curly arrow has a post on his blog where he's using a double oblique stopper to control the vacuum, with one of the ports in the stopper being a fine capillary. The reasoning being, a finer capillary will create a poorer connection with the vacuum pump.

That demonstrates that the vacuum pump is pulling an excessively hard vacuum in the first place. And, if you need a continuous flow from the pump, that also indicates that the trap is not actually trapping the solvent with a high efficiency.

If the trap is functioning, there should be zero flow through the vacuum pump - bar the leakage from the joints.

Also on the point of convenience, rotovaps tend to reduce the boiling point to the lowest possible, to effect a very rapid boil off rate at low bath temperatures.

That means dry ice or liquid nitrogen really needs to be involved.

At home, it may be less of an issue. You could use a weaker vacuum and a decent, none dewar, condenser. In fact, quite a few rotovaps use coil condensers.

Going back to curly arrow and the others, I can understand that the most convenient source of a decent vacuum in universities is usually coming from a rotary or a rotary with a diffusion pump on it.

However, if you connect most volatile solvents to the vacuum from a fridge compressor, the boiling point will drop to ten to a few tens of degrees C.

This is an important point because, without resorting to dry ice and liquid nitrogen, it is also the temperature range where water cooled condensers begin to fail in terms of liquefying.

It's usually recommended you have around 20C of temperature different between your condenser and the BP of the thing going through it. This changes based on the efficiency of the condenser and the flow rate of the thing you are putting through it. You can use liebigs with very little temperature difference, provided you spend hours doing it slowly.

In a rotovap, the goal is usually speed - as much solvent through as quickly as possible.

In general then, you want your new BP under vacuum to not be below 10C, and preferably more like 20C. A coil condenser is needed for any form of decent flow rate. You will struggle to keep a water condenser close to zero without resorting to buckets of ice or putting the entire thing in the freezer.

[Edited on 8-2-2011 by peach]




View user's profile View All Posts By User
DJF90
International Hazard
*****




Posts: 2266
Registered: 15-12-2007
Location: At the bench
Member Is Offline

Mood: No Mood

[*] posted on 8-2-2011 at 09:39


When using a rotovap it is actually advised to have the vacuum set so that the boiling point of the solvent is 40*C. This then allows for your 20*C difference with the condenser cooling water, and the bath itself should be set to 60*C - this is what Buchi teaches, according to a past member. And it seems to work fine. The issue most people will have is controlling the vacuum. If you can get a gauge on there somewhere, the air leak at the top of the condenser can be adjusted to give some crude control.

In reality, vacuum is still pulled with solvents like diethyl ether and dichloromethane, and as such these usually end up going through the pumps. There are commercial vacuumstations which utilise a PTFE diaphragm pump and a condenser assembly on the OUTLET, allowing for near 100% solvent recovery.

[Edited on 8-2-2011 by DJF90]
View user's profile View All Posts By User
peach
Bon Vivant
*****




Posts: 1428
Registered: 14-11-2008
Member Is Offline

Mood: No Mood

[*] posted on 8-2-2011 at 12:47


Using buckets of ice, I've seen the temperature of the water in my condenser a few degrees above zero. Buckets of ice are a lot of hassle, so I'll assume you're working straight off the tap. It's winter here in the UK and, having just stuck a thermometer under the tap, the water is coming out at 7-8C. Assuming a 20C difference for condensation, that's a BP of around 30. And, given that the tap water is going to warm up in summer, that's 40C.

Atmospheric pressure (1 ATM) is 760 Torr (mmHg), 1013.5 mBar and 101,350 Pascals.

Running the more common solvents through the pressure calculation for a BP of 40C gives;

Acetone
424 torr, 560 mBar

Ethyl Acetate
208 torr, 277 mBar

Toluene
51 torr, 67 mBar

Hexane
297 torr, 396 mBar

Cyclohexane
183 torr, 244 mBar

Chloroform
357 torr, 472 mBar

Diethyl Ether - Not even going to bother working that out, it boils at 35C under 1 ATM. If you're running ether through a tap water condenser, under vacuum, you should be immediately disqualified from the course.

DCM - Same, boils at 40C under 1 ATM

Isopropyl Alcohol
103 torr, 137 mBar

Ethanol
135 torr, 180 mBar

Methanol
258 torr, 344 mBar

Water
62 torr, 82 mBar

Here are some examples of the levels of vacuum generated by different methods;

A vacuum cleaner
800 mBar

Sucking with my mouth :P
200 mBar

Basic diaphragm pump (I wonder how aquarium pumps do? Anyone got one they want to post me to check?)
100 mBar

Water aspirator
32 mBar at 25C

The fridge pump I've left on since last week, after putting all kinds of stuff through it
29 mBar

Dual stage HVAC rotary vane
0.05 mBar

Dual stage laboratory rotary vane (the vacuum source attached to Schlenk lines)
0.001 mBar

Diffusion pump (also found on Schlenk lines)
0.000000001 mBar

As you can see.... both the rotary and diffusion pump are WAAAAAAAAY out the other end.

If I put the rotary pump pressure through for say, acetone, we get a BP of -124C.

Dry ice melts (sublimes) at -78C. No where near close enough, it's on the wrong side of the BP for a start.

If a diffusion pump is used, the BP of the acetone will drop to -176C. The only way to trap it is going to be liquid nitrogen, at -196C. Once you're using liquid nitrogen, you then have the potential of liquefying oxygen (-183C) in the condenser as well. Liquid oxygen + volatile solvent ---> Fire

Let's now take a look at the prices!

Vacuum cleaner
Free, steal it and pretend you're doing the tidying up. Downside, it's not strong enough and vacuum cleaner motors SPEW sparks out in the gas path

My mouth
Depends on the mood and how broke I am at the time. Downside, not a particularly controllable source of vacuum and I'll get a headache

Diaphragm pump
£880 for the BOC Edwards D-Lab (their most basic)

Aspirator
£20 for a new Chapman from eGimp

Fridge pump
Free. Or, if you want to be classy like MossyDie, £2.30 new from eBay.

Dual stage HVAC rotary vane
£73

Dual stage rotary vane
£1376.4 Edwards E2M0.7 (cheapest they do)

Diffusion pump
£692 Chemglass with mantle

SO!!!!!!!

You can see that for a 40C condensation point, you are looking at a diaphragm, fridge or aspirator pump for the majority of volatile solvents. A HVAC if you want it all covered.

Taking acetone as a model example, a dual stage rotary vane lab pump is five orders of magnitude too strong in terms of pressure and a diffusion is 11 orders of magnitude out.

Looking at price, and comparing it to MossyDies new fridge pump, at £2.30, the laboratory rotary is three orders of magnitude more expensive, and the diffusion is the same.

You also have to factor in for other issues. For example, if you use the laboratory rotary and diffusion, that means buying dry ice and liquid nitrogen, storing them, and having the equipment to handle them.

Putting things under vacuum dose not change their enthalpy of vaporisation. It doesn't make more boil off per unit of heat going in. The only thing that changes is the temperature gradient. Going back to acetone, you're looking at a gradient of about 250C (DIFFERENCE) if the flask is in a hot water bath. Borosilicate is rated to +400C. So you can see there is potential to put the flask under vacuum using one of the other methods, then heat it in an oil bath to establish the same gradient or more.

The gas in a system at 100 mBar is at a density of 10^17 molecules per cubic centimetre. Under a lab rotary it's 10^12, and under a diffusion it's 10^7. The density of the material going through the condenser has dropped many zeros.

The bad method

One Buchi R-210 rotovap
£4246

+ A BOC Edward E2M0.7 + Chemglass Diffusion + Chemglass Schlenk line

£8705

I mean FUCK! To boil a solvent off? And not condense it?

Conclusion

If condensing with tap water or a tank of water and a pump is the goal, it should be very apparent which are the smarter choices from the above numbers.

Tanks of water with a pump in them have a horrible tendency to warm up whilst you go for that quick cup of coffee. And if you're using ice slush, the pressures won't need to go much lower to meet up with those new BPs.

This, £8705 way of improving education, is why tuition fees have jumped from £3k to £9k this year. And it's improved it so much people, with PhD's, now think that is a good idea.

[Edited on 8-2-2011 by peach]




View user's profile View All Posts By User
Regolith
Hazard to Self
**




Posts: 73
Registered: 4-2-2011
Location: Mining the moon.
Member Is Offline

Mood: Glacial

[*] posted on 8-2-2011 at 23:55


I DO like Peach doing all the googling for me, I like to have actual boiling points not just memory to go by when composing my thoughts. I'm confused by the difference between the 2 devices I know what each does and it's intent but if you could only own ONE. The answer is an obvious full distillation setup with stir bar. The downside is it's gonna take a little longer, you can't go above the curie temperature of the magnetic stir bar and your wallet is gonna be way WAY heavier, like its gonna take some effort to haul around the extra money you didn't spend. Wait, did I say bad thing ?
View user's profile View All Posts By User
turd
National Hazard
****




Posts: 800
Registered: 5-3-2006
Member Is Offline

Mood: No Mood

[*] posted on 9-2-2011 at 00:54


Quote: Originally posted by peach  

Diffusion pump (also found on Schlenk lines)
0.000000001 mBar

No way you'll get 10^-9 mbar on a Schlenk line. :o
Quote:

The bad method

One Buchi R-210 rotovap
£4246

+ A BOC Edward E2M0.7 + Chemglass Diffusion + Chemglass Schlenk line

£8705

I mean FUCK! To boil a solvent off? And not condense it?

All roties I have seen were connected to controlled(!) diaphragm pumps like this one:

They're too pricey for the amateur, but a research institution will buy them by the truck load (literally!), making them much less expensive. Also, in the situation you describe the rotary pump is needed for the Schlenk anyway.

Concerning tuition fees, complain to your politicians. :D
View user's profile View All Posts By User
madcedar
Hazard to Others
***




Posts: 116
Registered: 10-9-2009
Member Is Offline

Mood: No Mood

[*] posted on 9-2-2011 at 01:11


Another article on rotary evaporators here: http://www.cookingissues.com/primers/rotovap/
View user's profile View All Posts By User

  Go To Top