Sciencemadness Discussion Board
Not logged in [Login ]
Go To Bottom

Printable Version  
Author: Subject: Soxhlet
nimgoldman
Hazard to Others
***




Posts: 303
Registered: 11-6-2018
Member Is Offline


[*] posted on 8-8-2018 at 22:20
Soxhlet


I am using Soxhlet apparatus quite often for making plant extracts.

My 250 ml Soxhlet body became too small for low yielding plants and hot swapping thimbles was too cumbersome so I decided to purchase a 1000 ml Soxhlet body (see picture).

Unfortunately, it has the drainage tube U turn levelled almost at the half of the body (marked by red line in the picture), meaning that the solvent will drain before even getting to the top.

This means I cannot actually load the Soxhlet to its full capacity and I am still left with almost the same usable volume as with my older 250 ml Soxhlet.

Can you please explain me why the drainage tube is positioned so low, rendering most of the Soxhlet volume pratically unusable?

soxhlet-issue.jpg - 23kB
View user's profile View All Posts By User
Vosoryx
Hazard to Others
***




Posts: 282
Registered: 18-6-2017
Location: British Columbia, Canada
Member Is Offline

Mood: Serial Apple Enjoyer

[*] posted on 8-8-2018 at 22:28


I have wondered the same thing often. It seems to make no sense, unless for using less solvent.
But if that were the case, then it seems like that would be a specialty item. (with the regular item being the one with the siphon tube going to the top) Usually, the solvent is what you have in abundance and the solute is not.

I have no idea but i'm curious if anyone knows.

Great question!




"Open your mind son, before someone opens it for you." - Dr. Walter Bishop
View user's profile View All Posts By User
Deathunter88
National Hazard
****




Posts: 507
Registered: 20-2-2015
Location: Beijing, China
Member Is Offline

Mood: No Mood

[*] posted on 9-8-2018 at 01:11


I would assume it's because most of the efficiency of a Soxhlet comes from the hot vapors that wash over whatever you're extracting and condensing onto them. Therefore, even if your plants are not soaking in liquid solvent, they are still being extracted. After all, the point of a Soxhlet is not to soak leaves in solvent (you could just dump them into a beaker for that), but rather to circulate fresh solvent, and in that regard the siphon tube height has little effect on efficiency.

Again, this is speculations, but I believe it to be truthful.
View user's profile View All Posts By User
Vosoryx
Hazard to Others
***




Posts: 282
Registered: 18-6-2017
Location: British Columbia, Canada
Member Is Offline

Mood: Serial Apple Enjoyer

[*] posted on 9-8-2018 at 07:44


I thought about that explanation, but it doesn't make sense to me.

If hot vapour were what did the extraction, why bother with a siphon tube at all? Seems like a way to make it very complicated for no reason.

Maybe. I just don't see why.




"Open your mind son, before someone opens it for you." - Dr. Walter Bishop
View user's profile View All Posts By User
Fulmen
International Hazard
*****




Posts: 1693
Registered: 24-9-2005
Member Is Offline

Mood: Bored

[*] posted on 9-8-2018 at 10:29


No, it does not work like that. The hot vapors will not be in much contact with the sample, and with a water cooler the condensate will probably be fairly cold.
It works by having a constant stream of pure solvent running through the sample. So the sample doesn't have to be fully submerged as long as the condensate drips down into the thimble.
In fact allowing the thimble to be (almost) fully submerged could be risky. If it clogs up you risk washing particulates over the top, contaminating the extract or even clogging the siphon.




We're not banging rocks together here. We know how to put a man back together.
View user's profile View All Posts By User
Abromination
Hazard to Others
***




Posts: 432
Registered: 10-7-2018
Location: Alaska
Member Is Offline

Mood: 1,4 tar

[*] posted on 9-8-2018 at 20:28


Vosoryx, how expensive was your soxhlet?
Im considering buying one.




List of materials made by ScienceMadness.org users:
https://docs.google.com/spreadsheets/d/1nmJ8uq-h4IkXPxD5svnT...
--------------------------------
Elements Collected: H, Li, B, C, N, O, Mg, Al, Si, P, S, Fe, Ni, Cu, Zn, Ag, I, Au, Pb, Bi, Am
Last Acquired: B
Next: Na
--------------
View user's profile View All Posts By User
Vosoryx
Hazard to Others
***




Posts: 282
Registered: 18-6-2017
Location: British Columbia, Canada
Member Is Offline

Mood: Serial Apple Enjoyer

[*] posted on 10-8-2018 at 12:13


Quote: Originally posted by Abromination  
Vosoryx, how expensive was your soxhlet?
Im considering buying one.


Don't remember to be honest.
I got it on eBay, from deschem. They're pretty cheap and the quality is good.
Amazing customer service too.




"Open your mind son, before someone opens it for you." - Dr. Walter Bishop
View user's profile View All Posts By User
CharlieA
National Hazard
****




Posts: 645
Registered: 11-8-2015
Location: Missouri, USA
Member Is Offline

Mood: No Mood

[*] posted on 10-8-2018 at 16:28


I'm inclined to agree with Fulmen.
I would think the top of the siphon tube could be a little higher. A plug of cotton or glass wool at the top of the thimble might also prevent the solids from getting "loose" and plugging up the siphon.

Other than time or a low content of the desired material in the raw material, I would think that the smaller Soxhlet would work as well as the larger. If solubility of the product in the solvent/receiver flask were an issue, could this be solved by going to (and thereby using more solvent)?
View user's profile View All Posts By User
Mr. Rogers
Hazard to Others
***




Posts: 184
Registered: 30-10-2017
Location: Ammonia Avenue
Member Is Offline

Mood: No Mood

[*] posted on 11-8-2018 at 02:58


If the siphon tube is lower down in the body the solvent will recirculate faster = You need to use less. That's all it is.


[Edited on 11-8-2018 by Mr. Rogers]
View user's profile View All Posts By User
unionised
International Hazard
*****




Posts: 5102
Registered: 1-11-2003
Location: UK
Member Is Offline

Mood: No Mood

[*] posted on 11-8-2018 at 03:32


The sample in the thimble has a constant stream of clean hot solvent dripping onto it.
That's going to do the extraction slightly better than sitting in a thimble full of solvent.
View user's profile View All Posts By User
Yamato71
Hazard to Self
**




Posts: 68
Registered: 8-2-2012
Member Is Offline

Mood: Resigned

[*] posted on 18-8-2018 at 08:10


You have to remember that the extraction chamber has more than solvent in it. There is also the thimble and its contents, which displaces solvent. You don't want the solvent level to rise above the mid point of the thimble for two very good reasons. First, you want the refluxed solvent to travel through as much of the column being extracted as possible. if the liquid level is allowed to rise to the upper part of the thimble, the portion below the liquid level isn't being efficiently extracted because it's steeping in stagnant solvent. Once the solvent permeating the column reaches the liquid level, it stops flowing downward. By discharging pregnant solvent early and often, almost all of the column material is extracted by flowing solvent.
The second issue created by a high solvent level is contamination. Should the solvent level top the thimble, bits of plant matter will float out of the thimble and find their way into the stillpot. Even worse, plant matter can clog the thin delicate siphon tube, trapping solvent in the upper regions of the setup and allowing the boiling flask to boil dry.

[Edited on 18-8-2018 by Yamato71]
View user's profile View All Posts By User

  Go To Top