Sciencemadness Discussion Board
Not logged in [Login ]
Go To Bottom

Printable Version  
Author: Subject: Brand new hotplate won't boil water?
Chemist1357
Harmless
*




Posts: 29
Registered: 26-10-2022
Member Is Offline


sad.gif posted on 20-3-2023 at 04:25
Brand new hotplate won't boil water?


After spending $300 I'm furious to say the least. The window to return the product has lapsed. The brand is "Four E's Scientific" and it's rated to go 280c. I put the probe in 200ml of water and after 1 hour the hotplate won't go above 96c. I have the temperature cranked all the way up to 280c and it still won't go above 96c. My infrared thermometer says the plate is 220c+ but the water won't get past 96c.

Is there anything I can do to fix this? I just read reviews and it seems to be a common thing. If I can't fix this can anyone recommend a good hotplate/stirrer with a temperature control probe that is NOT made in China.



Hotplate.JPG - 2.1MB

[Edited on 20-3-2023 by Chemist1357]
View user's profile View All Posts By User
blogfast25
International Hazard
*****




Posts: 10562
Registered: 3-2-2008
Location: Neverland
Member Is Offline

Mood: No Mood

[*] posted on 20-3-2023 at 05:59


First and foremost, check BOTH your thermocouple and IR thermometer, the discrepancy is oddly great.

Secondly, check temperature setting on device.

Thirdly, contact vendor/manufacturer anyway, you never know...




View user's profile View All Posts By User
Rainwater
National Hazard
****




Posts: 993
Registered: 22-12-2021
Member Is Offline

Mood: Break'n glass & kick'n a's

[*] posted on 21-3-2023 at 01:22


Check the manual
https://images-na.ssl-images-amazon.com/images/I/91UOqMS+OuL...

Page 9
There is a upper tempature limit setting.
A screw next to the power switch.




"You can't do that" - challenge accepted
View user's profile View All Posts By User
blogfast25
International Hazard
*****




Posts: 10562
Registered: 3-2-2008
Location: Neverland
Member Is Offline

Mood: No Mood

[*] posted on 21-3-2023 at 05:03


Quote: Originally posted by Rainwater  

Page 9
There is a upper tempature limit setting.
A screw next to the power switch.


Well spotted!

Turns out that my own recently bought heater/stirrer also has a max temperature safety that I didn't know of! :o:o

[Edited on 21-3-2023 by blogfast25]




View user's profile View All Posts By User
Dr.Bob
International Hazard
*****




Posts: 2918
Registered: 26-1-2011
Location: USA - NC
Member Is Offline

Mood: Mildly disgruntled scientist

[*] posted on 21-3-2023 at 13:12


The other main issue is how many watts does it consume at max? many good hotplates are 600+ watts, but the cheap ones tend to be 200-400 watts, and that won;t heat much other then the top of the hotplate to a high temperature. I would go with a used Corning hotplate any day over any "new"Chinese one. The Corning 320 or 420 models are both great, and I use them both daily, and all of mine are over 12 years old.
View user's profile View All Posts By User
BromicAcid
International Hazard
*****




Posts: 3323
Registered: 13-7-2003
Location: Wisconsin
Member Is Offline

Mood: Rock n' Roll

[*] posted on 21-3-2023 at 15:27


Also, even though you paid a good chunk of money for it. Precision can be expensive, it might not heat as well as you want but it might be designed to hold a specific temperature range well or to adjust fast or even interface with some external piece of equipment. When I was doing chemistry in my back yard I had one long thermometer, as long as any other that only covered from -10°C to 0°C (I won it in a lot) and I always thought that was odd till I got in the industry and realized that precision costs money.




Shamelessly plugging my attempts at writing fiction: http://www.robvincent.org
View user's profile Visit user's homepage View All Posts By User
Chemist1357
Harmless
*




Posts: 29
Registered: 26-10-2022
Member Is Offline


[*] posted on 22-3-2023 at 08:44


Quote: Originally posted by Rainwater  
Check the manual
https://images-na.ssl-images-amazon.com/images/I/91UOqMS+OuL...

Page 9
There is a upper tempature limit setting.
A screw next to the power switch.


I tried that, made sure it was all the way up to 320c.

Still didn't do anything.

Looks like it's new hotplate time.
View user's profile View All Posts By User
Texium
Administrator
********




Posts: 4745
Registered: 11-1-2014
Location: Salt Lake City
Member Is Offline

Mood: Seeking gainful employment

[*] posted on 22-3-2023 at 10:00


Quote: Originally posted by BromicAcid  
When I was doing chemistry in my back yard I had one long thermometer, as long as any other that only covered from -10°C to 0°C (I won it in a lot) and I always thought that was odd till I got in the industry and realized that precision costs money.
The perfect tool for when you need a very precise read on the temperature of your salted ice bath! :D



Come check out the Official Sciencemadness Wiki
They're not really active right now, but here's my YouTube channel and my blog.
View user's profile Visit user's homepage View All Posts By User
Rainwater
National Hazard
****




Posts: 993
Registered: 22-12-2021
Member Is Offline

Mood: Break'n glass & kick'n a's

[*] posted on 22-3-2023 at 16:19


If i was in your shoes, I'd rip it apart, ohm out the heating element.
Calculate total wattage. Or just read the lable and see how many amps ot pulls. But thats no fun

Determine how it's powered (pwm, triac, relay). All this can be determined with an amp probe without disassembling anything, again. No Fun

So, for every ml of water, it needs 4.19 watts of power to raise the temperature 1 degree C (without changing states)
To boil 1000ml ( 0c liquid ~ 100c liquid), you need to introduce 419000 watt of power ( watts = joules)
plus your total thermal losses.

For 30 minutes, that 233-watt heater
20 minutes, 350-watts
10 minutes, 700-watts.

All UL rated heaters have a thermal limit, which provents the device from exceding a set temperature.

Let's say your wattage is sufficient and the hotplat is reaching the maximum temperature its rated for, but your water is still not boiling.
The 3 factors I have had to fix myself are thermal losse, thermal conductivity, pid settings.

Is the bottom of the beaker perfectly flat?
Is the hotplate perfectly flat?
Is the hotplate turning off more as it approaches the set point?

Quick test for good contact. With the appratus cool, wet the bottom of an empty beaker, and place it onto the plate.
Watch the water pattern as you set the beaker down. Notice any voids, puddles, bibbles. If you see them, this is likely a problem
A few layers of aluminum foil under the beaker can fix this.

Thermal losses can be beat by using an insulated wrap that is compatability with the reagents your using.

Pid setting can be tested by repeating your above described experiment with the pt1000 sencer located laying on the table top. And at room temperature.
This will cause the controller logic to keep applying heat.

And remember, if hitting it with a hammer doesnt fix it, hit it harder to get a new one
Good luck




"You can't do that" - challenge accepted
View user's profile View All Posts By User
Fyndium
International Hazard
*****




Posts: 1192
Registered: 12-7-2020
Location: Not in USA
Member Is Offline


[*] posted on 22-3-2023 at 23:17


Do not buy Chinese hotplates unless a specific recommendation for a specific brand or type is given. Most of them are garbage and have little to no other than superficial heating power and the stirrer units mostly suck. They can get to 400-600C - when you measure it at the heater element, but they are incapable of heating anything smaller than few dozens of ml at most. Hotplates are generally good for smaller volumes only anyway, but these are next level garbage.

Yes, I paid for a lesson a decade ago and had the exact same issue. It was barely able to simmer a few hundred mL of water after an hour or two of heating.

Some heating mantles, at least those with good reviews, seem to be better, but they aren't neither professional quality.
View user's profile View All Posts By User
PirateDocBrown
National Hazard
****




Posts: 570
Registered: 27-11-2016
Location: Minnesota
Member Is Offline

Mood: No Mood

[*] posted on 23-3-2023 at 03:00


I suspect these Chinese ones are made for any market, and when they sell to countries with 110V service, they just install a power cord for that, and ship it. The element and controls are still for 220V.

Since you have nothing to lose, try a step up transformer, and test it on double voltage.




Phlogiston manufacturer/supplier.

For all your phlogiston needs.
View user's profile View All Posts By User
Rainwater
National Hazard
****




Posts: 993
Registered: 22-12-2021
Member Is Offline

Mood: Break'n glass & kick'n a's

[*] posted on 23-3-2023 at 07:10


Very good point. For multiple operation a 2 element design is comminly used. For 120v the heating elements should wired in parallel, 240v they will be in series.



"You can't do that" - challenge accepted
View user's profile View All Posts By User
Cathoderay
Hazard to Self
**




Posts: 97
Registered: 29-1-2023
Location: US-Texas
Member Is Offline


[*] posted on 23-3-2023 at 07:33


Quote: Originally posted by PirateDocBrown  
I suspect these Chinese ones are made for any market, and when they sell to countries with 110V service, they just install a power cord for that, and ship it. The element and controls are still for 220V.

Since you have nothing to lose, try a step up transformer, and test it on double voltage.

I wouldn't do that. If your suspicion is wrong you could damage the control circuits.
If it was just the heating element then it might not be as bad.
Many times there are internal connections that are changed to accommodate the different line voltage.
It is odd that the infrared thermometer says 220 C but the water in the beaker won't boil. If the plate was that hot you should be able to put some drops of water directly on the plate and it should boil away very fast.
As Rainwater said you might have a problem with contact or heat loss.
View user's profile View All Posts By User
Cathoderay
Hazard to Self
**




Posts: 97
Registered: 29-1-2023
Location: US-Texas
Member Is Offline


[*] posted on 23-3-2023 at 07:59


Also I think that many people put too much stock in those infrared thermometers.
Among other things the emissivity setting has to be correct for accuracy.
https://www.thermoworks.com/infrared-thermometry101-limitati...
View user's profile View All Posts By User
Sulaiman
International Hazard
*****




Posts: 3943
Registered: 8-2-2015
Member Is Offline


[*] posted on 23-3-2023 at 14:10


No mention yet of the rated power of the hotplate,
which should have been a very important part of the selection process.

This and similar lower powered hotplate stirrers are only intended to keep liquids at a near constant warm temperature, and stirred.

For cost reasons, I doubt that there is a dual heating element.

As mentioned above, measuring the resistance of the heating element will give an actual power rating.
eg W = V2/R (V = supply voltage)

I've never owned a beaker with a completely flat bottom - are they common ?




CAUTION : Hobby Chemist, not Professional or even Amateur
View user's profile View All Posts By User
Rainwater
National Hazard
****




Posts: 993
Registered: 22-12-2021
Member Is Offline

Mood: Break'n glass & kick'n a's

[*] posted on 23-3-2023 at 16:29


In my collection. No i put aluminum under everything.
Manual says 600watts 120/240v. An amp probe will reveal the accuracy of that manual
watts / volts = amps
120 = 5 amps
240 = 2.5 amps

Actually from the specs, it looks like a good one
24 month warranty.
600w heater
55mm × 10mm stirrer magnet
Only thing im not liking is the required use of an external pt1000 probe.
I would cutin a switch and install an additional sensor onto the underside of the hotplate so the external probe would be optional




"You can't do that" - challenge accepted
View user's profile View All Posts By User
Chemist1357
Harmless
*




Posts: 29
Registered: 26-10-2022
Member Is Offline


[*] posted on 24-3-2023 at 02:37


The voltage where I live is 220-240. We use the same plug as China. I did put aluminium foil underneath and it did absolutely nothing. I have a Chinese heating mantle that blasts through heat so it just seems this one is junk.
View user's profile View All Posts By User
Sulaiman
International Hazard
*****




Posts: 3943
Registered: 8-2-2015
Member Is Offline


[*] posted on 24-3-2023 at 04:44


Have you raised this issue with the seller yet?

I would expect a full refund.




CAUTION : Hobby Chemist, not Professional or even Amateur
View user's profile View All Posts By User
Cathoderay
Hazard to Self
**




Posts: 97
Registered: 29-1-2023
Location: US-Texas
Member Is Offline


[*] posted on 24-3-2023 at 07:48


I don't have one of those Chinese hot plates, however from various sources of info I think there is a sensor in the hotplate that is selected if you do not plug in the external probe.
From watching some NurdRage videos, some have two sensors under the plate, one controls the temperature and another is attached to the over-temperature safety limit. NurdRage had a hotplate were the temperature sensor failed and he was able to replace it. The display would continuously show the maximum possible temperature but the plate was not heated at all.
https://www.youtube.com/watch?v=Xa2OMNaHn_Y
I think the thermal design of those Chinese hotplates is not very good. The heating element itself may reach the correct temperature (that is where the sensors are) but the heat does not make it to the metal plate very well.
I doubt very much the plate can be at 280 C and the water isn't boiling. You would need to have an insulating blanket between the plate and the beaker.
View user's profile View All Posts By User
Rainwater
National Hazard
****




Posts: 993
Registered: 22-12-2021
Member Is Offline

Mood: Break'n glass & kick'n a's

[*] posted on 24-3-2023 at 14:27


The manual says do not operate with out the external sensor. If return/refund is no option, its time to go unpluged and dive deep into a crappy fake to see if its useable. My guess would be a simple defect.
Things to look for

Missing/burnt components
Lose connections
How well the heater and hotplate are joined, is there thermal paste. Pay close attention to this part
Thermal limit sensor likely to be a pt1000 do to the external probe, (same hardware can read both)

There will be a minimum of 2 wires going to the heater connected to the board, ohm them out.
If your qualified, plug up the unit and check the voltage going to the element. Dont get killed.
And pictures, top down, text and color bands in focus.

These things are usually very easy to fix do to the unit, kinda working.
Ive gotten full refunds on a lot of equipment that hasnt worked and is to expensive to ship back
managed to fix most of it.




"You can't do that" - challenge accepted
View user's profile View All Posts By User
Chemist1357
Harmless
*




Posts: 29
Registered: 26-10-2022
Member Is Offline


[*] posted on 28-3-2023 at 22:54


Quote: Originally posted by Sulaiman  
Have you raised this issue with the seller yet?

I would expect a full refund.


I would have but I bought it months ago and never used it so it’s passed the window to return it
View user's profile View All Posts By User
Sulaiman
International Hazard
*****




Posts: 3943
Registered: 8-2-2015
Member Is Offline


[*] posted on 29-3-2023 at 03:46


I have quite a lot of free stuff from China due to claiming refunds,
sellers rarely want to pay return shipping so I get partial distillation kits, often repairable electronics etc.
I used to bargain for fair compensation with sellers
but I got so much hassle that nowadays
I test on arrival and claim in full from the seller platform if any problem.




CAUTION : Hobby Chemist, not Professional or even Amateur
View user's profile View All Posts By User
Texium
Administrator
Thread Moved
5-12-2023 at 14:05
JackleApp
Harmless
*




Posts: 7
Registered: 29-4-2025
Member Is Offline

Mood: No Mood

[*] posted on 30-4-2025 at 17:06


I got an 85-2, from the photos it looked pretty big and clunky, and it was on special, so got that as my first stirring plate, didnt notice the 100W power rating spec. this thing barely heats coffee, and indeed is all i use it for now, esp since its forced in closed loop mode so you cant heat anything that the low grade stainless probe isnt compatible with, which is just about anything
View user's profile View All Posts By User

  Go To Top